AGFA CR 30X Cassettes not Detected ERROR

Experiencing a Problem when scanning 35X43 Cassettes , Giving a 4152 ERROR (unknown cassette format detected) while small cassettes are working fine , did stall calibration but it seems everything is ok . any advices are highly appreciated.

 

Kind regards,

Muhammed


   RE: Computed Radiography Report Post
Sun Nov 06 2016Reply from Helper from north
avatar placemark

Here are what Agfa sayes

Error 4152

Meaning:
The MOTOR CASSETTE FIXING M1 detected an unknown cassette format. The cassette is still clamped and a user message is displayed to press confirm to release the cassette. The digitizer is still working, but if this error occurs frequently, it requires service support. Reason:
The MOTOR CASSETTE FIXING M1 measured a cassette width which doesn't match with a known cassette format. The determined cassette width is too small or too big due to:
- The cassette was inserted wrongly by the user (skewed, upside down or not with the flap first)
- The CASSETTE FIXATION STEEL BAND is not moving smoothly
- The CASSETTE DETECTION HOOK is not properly fixed or not the improved version
- Load detection for MOTOR CASSETTE FIXING M1 is imperfect, STALL CALIBRATION was not successfully
- Drive mechanism of MOTOR CASSETTE FIXING M1 is blocked or defect Cure:
1. Remove the cassette and insert it correctly again.
2. Check if cassette can be pushed smoothly into the right edge of the input slot and that the wedge of the cassette fixer presses correctly into the left notch of the cassette.
3. Insert another cassette of the same (preferred) or any other size.
4. Check if the CASSETTE  FIXATION STEEL BAND and the cassette fixing motor is moving smoothly.
5. Check if the CASSETTE DETECTION HOOK is properly fixed. If applicable mount the improved CASSETTE DETECTION HOOK, refer to Service Bulletin No.35
6. Perform a calibration of the MOTOR CASSETTE FIXING M1, refer to the Service Manual
7. Replace the MOTOR CASSETTE FIXING M1, refer to the Service Manual for order and replacement
8. Replace the IP HANDLING BOARD, refer to the Service Manual for order and replacement
9. Replace the complete CASSETTE FIXATION UNIT, refer to the Service Manual for order and replacement

 


Report Post
Mon Jan 21 2019Reply from S. Popov
avatar placemark

I have the same problem and I think the problem is from CASSETE DETECTION HOOK(have cracks).

What about this Service Bulletin No.35 ? Does it have this improved CASSETTE DETECTION HOOK like spare part number, or only whole Fixer Unit???


Report Post
Wed Aug 25 2021Reply from saliou87
avatar placemark

Good evening following a problem with my HP computer that was sold to me with an AGFA CR 30-X device (ABC-CODE: 5RB65 001, whose Product is: NX8500 SU1 HP8000 RAID1 PC the TYPE application is: PROD with an NX_X.0.8500 SU1 installation which has the serial number: NX279-10053) that I can no longer work because of my computer and the NX Office Viewer software. I am writing this letter to you in order to find a solution to this so that I can continue my work quietly.