When we switch on the autoclave (Mocom Millennium B) we first get

"r. E4002/BM405021 S/N 12BMO294" followed by "LOCKING PROBLEM ALARM". It's impossible to continue after theses messages. What can we do?

Thank you, Manuela


   RE: Sterilizers/Autoclaves Report Post
Sat Nov 12 2016Reply from moumen
avatar placemark
The door it's open ??

Report Post
Sat Nov 12 2016Reply from moumen
avatar placemark
Send to me the err (take picture) moumenimhest@gmail.com

Report Post
Mon Nov 14 2016Reply from seneque5
seneque5

In locking problem there is 3 alarms: A022/ A023/A024

Do you have one of these three alarms?

 



Abdelmajid Bougrine
Report Post
Mon Nov 14 2016Reply from Manuela
avatar placemark

Yes, at first it was A022 and than changed to A023. Can you help me with this? Thanks


Report Post
Mon Nov 14 2016Reply from Manuela
avatar placemark

Quote moumen

Send to me the err (take picture) moumenimhest@gmail.com
 

 

The error appears when the door is both closed and when its opened. I'm sending you the email of the error. Thank you in advance

 

 


Report Post
Mon Nov 14 2016Reply from seneque5
seneque5

Yes I can, in 3 hours 4 PM GMT



Abdelmajid Bougrine
Report Post
Mon Nov 14 2016Reply from seneque5
seneque5

When the alarm A022 occurs, Do you interrupt the cycle, or due to black-out?

Passed from A022 to A023 means that at the switching ON with door closed, the micro switch was blocked! when turning the alarm is A023 the two micro switch are off!

1st suggestion: Black-out

2nd burned fuse

3th power supply problem



Abdelmajid Bougrine