This time we have a Ligasure electrosurgery unit, with error E115.
Is there anybody having the service manuals of this unit or can help with
this error?
will appreciate any help.
With many regards form Greece
John
Sun Jul 19 2009
Ligasure electrosurgery Error E115
Sun Jul 19 2009Reply from Heathers
John,
The service manual states that error number 115 is a "Master microprocessor
reported a powerup DOSERR test failure" (MM_DOSERR_TEST_ERROR).
Recommended action is to cycle power, if the error reappears, replace the
control board. If the error reappears, replace highvoltage power supply. If
the error reappears, replace RF output FET Q5 on the RF board.
Hope this helps.
Tim
Report Post
Fri Aug 13 2010Reply from colemc
Hi. Could you email me a copy of the LigaSure Service Manual please? I would much appreciate it. cole_mcallister@comcast.net
Thank You!
Report Post
Sat May 04 2013Reply from ansarimedical
Wed Aug 30 2017Reply from rom
Hi, I got LigaSure generator for repair, he show error 145 and 115, Q5 is ok, and HV board is ok too,what i can do with control board for check?