Ligasure electrosurgery Error E115

Hello dear list friends, This time we have a Ligasure electrosurgery unit, with error E115. Is there anybody having the service manuals of this unit or can help with this error? will appreciate any help. With many regards form Greece John

   RE: Electrosurgical Unit (ESU) Report Post
Sun Jul 19 2009Reply from Heathers
avatar placemark
John, The service manual states that error number 115 is a "Master microprocessor reported a powerup DOSERR test failure" (MM_DOSERR_TEST_ERROR). Recommended action is to cycle power, if the error reappears, replace the control board. If the error reappears, replace highvoltage power supply. If the error reappears, replace RF output FET Q5 on the RF board. Hope this helps. Tim

Report Post
Fri Aug 13 2010Reply from colemc
avatar placemark
Hi. Could you email me a copy of the LigaSure Service Manual please? I would much appreciate it. cole_mcallister@comcast.net Thank You!

Report Post
Sat May 04 2013Reply from ansarimedical
avatar placemark

can you please email me the service manual too.

 

Thank you,

Atif

attifsaleem@gmail.com


Report Post
Wed Aug 30 2017Reply from rom
avatar placemark
Hi, I got LigaSure generator for repair, he show error 145 and 115, Q5 is ok, and HV board is ok too,what i can do with control board for check?