Mon Nov 28 2016

Angle Indicator is not accurate

have this totalcare bed, problem is When i Press the Bed Flat control Botton the bed suppose to become leveled but the angle indicator is bettween 0 and 5  instead of 0.  same as trendelenburg, when i press revirese trendelenburg or vise versa the bed suppose to return to flat position and should stop but still angel indicator is between 0 and 5 instead of 0.  dose anyone know what the issue might me?  thanks


   RE: Bed Report Post
Mon Nov 28 2016Reply from Joe Sosnowski
avatar placemark

step 1: Home pg 2--> Status--> check "trend angle" to confirm bed thinks its at 0 degrees trend.

step 2: if the bed thinks its at 0 degrees, but you know it really isn't, recalibrate the hilow sensors from the service menu.

 

Good luck,

Joe


Report Post
Tue Nov 29 2016Reply from Nikou2x2
avatar placemark

I checked the Trend Angel and yeah its 0 degrees, but the angle indicator is between 0 and 5 degrees, so im guessing i have to recalibrate the scale?  I tried to find the Service manual for the Total care, but no luck.  Managed to find a service manual for a VersaCare that shows how to recalibrate, is it kinda the same procedures or diffrent?  does anyone have the manual for the TotalCare? I've replaced valves and other stuff before, havne't done a calibration, don't think i'll be that hard lol any ideas???


Report Post
Tue Nov 29 2016Reply from MedWrench MM
MedWrench MM

If you are in need of immediate assistance please feel free to reach out directly to our Beds sponsors. They are happy to help.

DiaMedical USA

MedMattress

Alco Sales & Service

 

Just be sure to tell them you were referred by MedWrench.


Jonathan F. Payne,
Media Manager @ MedWrench,
866-989-7057,
Jpayne@MedWrench.com
Report Post
Tue Nov 29 2016Reply from Joe Sosnowski
avatar placemark

You'll need to recalibrate the hilow sensors:

Enter the service menu --> sensor calibraion--> Run hilow all the way up, take a reading --> Run hilow all the way down, take a reading --> save and exit.

Should be good to go.


Report Post
Tue Nov 29 2016Reply from Nikou2x2
avatar placemark

Thanks for the help man, seems pretty easy, but i dont have these options in the main menu, All i see are Scale Functions, Config.Bed exit alarm, Bed Setup/Reset, Preset bed positions and lastley Status.  The bed that i have is a TotalCare Model... Any idea how i could get to Service menue----Sensor----Calibration??


Report Post
Tue Nov 29 2016Reply from Joe Sosnowski
avatar placemark

Yep, if its the GCI with the up and down arrow keys, you hold down the up and down arrow keys simultaniously until it enters the service mode.

If its a touchscreen GUI, you press Tools--> Enter Service Mode--> 812


Report Post
Wed Nov 30 2016Reply from Nikou2x2
avatar placemark

thanks man, finally found it, had to hold both up and down arrows...did hilow all the way up and hilow all the way down took a reading and saved and exited. didnt solve the problem, the angle indicator is still not at zero, also when i did hilow down the bed started beeping and service light started to blink orange color, i unpluged the bed and it stopped beeping and now the service light is blinking and angel indicator is still now at 0 degreess, this bed is pain in the rear now... any ideas??


Report Post
Wed Nov 30 2016Reply from Nikou2x2
avatar placemark

little update, just cheked the bed and the service light if off..  it has something to do with Trendelenburg position, Bed flat control wont return the bed to zero degree position, i do reverse trendelenburn to adjust the angel idicator to zero manually, so something is off, im gussing could be somthing inside side rail faulty or trendelenburg hydrolics or could be a valve..


Report Post