Fri Dec 02 2016

Fill light is not coming on and no water entering the chamber

Fri Dec 02 2016Reply from MedWrench MM
MedWrench MM

If you are in need of immediate assistance please feel free to reach out directly to our Sterilizers/Autoclaves sponsor RPI. They are happy to help.

Just be sure to tell them you were referred by MedWrench.


Jonathan F. Payne,
Media Manager @ MedWrench,
866-989-7057,
Jpayne@MedWrench.com
Report Post
Fri Dec 02 2016Reply from Cbet_Matt
Cbet_Matt
Check with Midmark tech support they are helpfull. RPI tech support is a good resource too


Matt

 
Report Post
Mon Dec 05 2016Reply from Walker0500
avatar placemark

And there's always Duraline! We'll be happy to help you out in any way we can. 

Duraline Systems Inc.

877-561-0500

www.duralinesystems.com

customercare@duralinesystems.com

 


Report Post
Mon Dec 05 2016Reply from MedWrench MM
MedWrench MM

Duraline is also happy to help as well. :)


Jonathan F. Payne,
Media Manager @ MedWrench,
866-989-7057,
Jpayne@MedWrench.com
Report Post
Fri Jan 27 2017Reply from SIU2014
avatar placemark

i would do a fill valve check to make sure its working properly. you can do this by going into the diagnostice menu.


Report Post
Fri Jan 27 2017Reply from Duraline Systems Inc.
avatar placemark

Please note that the dignostic menu is only on the new 2nd generation M11's. If it has a "fill light" this would tell me its the 1st gen. If you need any further help let me know.

Duraline Systems Inc.

877-561-0500

www.duralinesystems.com

customercare@duralinesystems.com


Duraline Systems Incorporated
39 Western Hwy. West Nyack, NY 10994
www.duralinesystems.com
877-561-0500

Report Post
Tue Jun 25 2019Reply from garbleduser
avatar placemark

 It’s likely that you have a short between the chamber water level sensor and the body of the autoclave. My suggestion is  with the unit unplugged, check continuity between the two.  If it’s low resistance in between the chamber and sensor in the back of the chamber it would tell the main board the chamber is already full.   Does it go straight into the heating portion of the cycle after start is pressed?  If so then this is likely your problem. There could be a fray in the wire going to the main board and touching a grounded surface outside the chamber inside the case of the unit,  or there may be a piece of conductive material bridging the inner chamber sensor to the chamber. 

 If the door is slightly ajar or the door closed switch is malfunctioning or disconnected the unit won’t fill either. 


Report Post
Tue Jun 25 2019Reply from Rix
avatar placemark

Common issue following cleaning. First thing for possible easy fix: ( be sure that there is room in the reservoir by checking the level is down at least two inches from max) place water in the CHAMBER manually until up to top of water dam. Then, with no contents , run a short/instrument cycle. Assuming normal completion, start another cycle and see if issue is resolved (ie, unit goes into fill mode for appropriate time and then heating for normal cycle. 


Report Post
Tue Jun 25 2019Reply from IndyWendy
avatar placemark

I would be happy to help.  I would start with cleaning the water level sensor.  If your still skipping the fill preform a manual fill, use the drain tube and fill the chamber.   

If the unit completes the cycle look at the water level sensor or the fill valve. 

 

If you are in need of a technican to come on site Edge provides coverage for the eastren half of the US.  

 

 


Wendy Argueta 
Edge Biomed 
wendy@edgebiomed.com
 

Report Post