will not reach 270 degrees

will not reach 270 degrees


   RE: Sterilizers/Autoclaves Report Post
Fri Dec 09 2016Reply from MedWrench MM
MedWrench MM

If you are in need of immediate assistance please feel free to reach out directly to our Sterilizers/Autoclaves sponsor RPI. They are happy to help.

Just be sure to tell them you were referred by MedWrench.


Jonathan F. Payne,
Media Manager @ MedWrench,
866-989-7057,
Jpayne@MedWrench.com
Report Post
Fri Dec 09 2016Reply from Ray Brown, CBET
avatar placemark

Steam supply low? Depending on where you are, man, it's cold out there today.. 

Ddo you have one of those "soaking" thermometers such as one from RPI? They are mercury thermometers that retain the highest temperature until they cool off and you sling them back down. They operate from 150F to 350F and RPI sells them for around $40. That will positively show you what your chamber temperature got up to. And generally the 270f is around 30 PSI, so if you are getting lower than 270 then you could have the supply issue, or maybe your safety valve is starting to release, prematurely, bleeding off a tiny bit of your pressure...


Report Post
Wed Dec 14 2016Reply from Kara Crissman
avatar placemark

Dear shadoe,

We came across your post about the AMSCO Century V116 Sterilizer. We have a dedicated Service Support team who can help. Please provide us with your preferred method of contact or call 1-800-333-8828 for assistance.

Sincerely,

Kara Daugherty
STERIS Corporation


Report Post
Thu Dec 15 2016Reply from GBlair
avatar placemark

It's been a number of years since I worked on those but I recall most issues were related to the exhaust manifold. Steris sells an exhaust manifold kit that will rebuild everything except the solenoid seats. If the seats are pooched, they sell a kit for that as well.


Report Post