Sun Jan 08 2017

Calibrate pressure and temperature

Calibrate pressure and temperature, diagrams, adjustments, measurements to be made on the boards.

thank you


   RE: Sterilizers/Autoclaves Report Post
Sun Jan 08 2017Reply from tjgdjn
avatar placemark
What is the problem you are having? What exactly are you needing to adjust?

Report Post
Sun Jan 08 2017Reply from Rigoberto
avatar placemark

Hello

The machine is over pressure, other times it does not reach the sterilization temperature, how and where to adjust with osciloscopio?

thank you


Report Post
Sun Jan 08 2017Reply from tjgdjn
avatar placemark
You may have a stuck solenoid. I don't know of any calibration procedures that can be done with an oscilloscope on this unit.

Report Post
Mon Jan 09 2017Reply from MedWrench MM
MedWrench MM

RPI may be able to help. Just give them a call and ask for Neil. Be sure to tell him you were referred by MedWrench.


Jonathan F. Payne,
Media Manager @ MedWrench,
866-989-7057,
Jpayne@MedWrench.com