Hello,

I have a Datascope Spectrum OR that defaulted back to German and would not hold the date/time. I tried to change it back to English by holding the freeze button while booting but it just repeatedly beeped at me and gave me a blank screen.

I changed the clock chip and it fixed the problem with it not holding time/date but it still repeatedly beeps when I try to access the installation menu to change the language. Does anybody know what could be causing this, or know how to change the language without this menu?


   RE: Vital Signs/Multi-Parameter Report Post
Wed Jan 18 2017Reply from ReNew14
avatar placemark

Sounds like the Main System PCB.

if you go into service mode and the display test fails instantly, then definitely Main PCB.

It's around 2500 for a new one.


Report Post
Wed Jan 18 2017Reply from lakeside wholesale medical
avatar placemark

Your Timekeeper/CPU monitoring chip may be bad. When you replace it on the Spectrum ORs it will reboot in Dutch and not allow you to change it. The cheap fix is to replace the chip and then send it to Mindray, and have them reload the software for around $700.00.

Thank you

Lakeside Wholesale Medical

Mike


Report Post
Fri Jul 28 2017Reply from Terrain Biomedical
Terrain Biomedical

Call Mark - he knows how to solve your issue!