The pump is giving a "warning replace battery" flashing message. I replaced battery with new battery.
The unit will not allow me to get into the Biomed setting (.)(start) to complete the change battery function. Any suggestions?
The pump is giving a "warning replace battery" flashing message. I replaced battery with new battery.
The unit will not allow me to get into the Biomed setting (.)(start) to complete the change battery function. Any suggestions?
Hi Mikeakchks,
If you are in need of immediate assistance please feel free to reach out directly to our Infusion Pump sponsors. They are happy to help out when they can.
Just be sure to tell them you were referred by Jonathan at MedWrench.
Two things: 1) Did you let the new battery charge for up to 8 hours? 2) When you turned on the pump and went to the [BIOMED SETTINGS] screen and pressed [CHANGE BATTERY], was the unit disconnected from AC power? - Cipher
Dear , Did you reset the battery?
If not do thses steps
1.switch of the machine .
2. on the unit and wait tell ask insert the set.
3.press ( . ) after that press start
it will apear battery in desplay press on it then it will be rest
4. switch the unit off .
5. after that on it it will be fine
I too have tried all the above but inspite of fully charging the new battery and following the above procedures the battery is only showing a partial charge. There was a beep when I hit hit the change button. But no change in charge level.