Fri Jan 20 2017

Secure II Bed Functional Test

I changed out the main circuit board and matcched the dip switches to the old board, but on pwer up the footboard shows "FuntionalTest".

I can't find the reason in the manual. Any Ideas? 


   RE: Bed Report Post
Fri Jan 20 2017Reply from Stryker Tech
avatar placemark
Recheck the dipswitches on SB4

Report Post
Fri Jan 20 2017Reply from MedWrench MM
MedWrench MM

Hi JATEC1,

If you are in need of immediate assistance please feel free to reach out directly to our Bed Sponsors. They are happy to help out when they can.

DiaMedical USA

MedMattress

 

Just be sure to tell them you were referred by Jonathan at MedWrench.


Jonathan F. Payne,
Media Manager @ MedWrench,
866-989-7057,
Jpayne@MedWrench.com
Report Post
Fri Jan 20 2017Reply from JATEC1
avatar placemark

Thanks Stryker Tech, it was exactly that. Problem solved!


Report Post