I have a table-top 6th gen. warming cabinet that is not heating up all the way. It heats to about 80-99 degrees. We've already replaced the control board and temp probe. Is it the relay? Thank you
Not heating up all the way
Yes - It's intermitent. We just went to look at it, and the set temp was 100 and it was at 99. We adjusted it to 115 and it rose to 115.
Report Post
it could be the relay. You need to verify if it is putting out voltage to the heater when it should be heating. you can do a resistance check on the heating element while tapping of flexing it a bit. Is the door good or is intermittantly opening? Does the door stay closed tightly? Does it alarm off overtemp?
Matt D.
Report Post
Dear emilymancini14,
We came across your post about the AMSCO Warming Cabinet. We have a dedicated Service Support team who can help. Please provide us with your preferred method of contact or call 1-800-333-8828 for assistance.
Thank you,
Kara Daugherty
STERIS Corporation
Report Post
IF you have repaired the unit. Please post what the problem was so others can look back on your problem and see if they can use your solution for their equipment problem.