I have an Alaris 8100 that continues to give the error "check IV set". I've replaced the AIL sensor and logic board, and it's still giving the error. Any suggestions?
"Check IV Set"
I agree
run the analog check and see if you can determine which sensor (fluid or patient side) is defective
anything over 2 volts for an unloaded sensor change it. Also it should change to 5volts or less as its pumping. If it stays at 5 volts or more-change it
Report Post
If you are in need of immediate assistance please feel free to reach out directly to our Infusion Pump sponsors. They are happy to help out when they can.
Just be sure to tell them you were referred by Jonathan at MedWrench.
Jonathan F. Payne,
Media Manager @ MedWrench,
866-989-7057,
Jpayne@MedWrench.com
Report Post
If you check your error log you will probably see one of the pressure sensors having an issue on and off in the past couple months. More than likely it is your patient side pressure sensor.
Report Post
if you connect to maintenance s/w you will se a selection in 8100 pump task list called "analog sensors" if you select this, you will see voltages for the bottle side and the Pt. side pressure sensors. it should be about 1.0 volt (probably slightly less, 0.9v) without the tube set installed and about 2-3v with tube set installed. way low (<0.5v) or way high (3.5-4.9v), indicates problem with that sensor. replacing sensor should correct problem. be aware a new sensor is about $105.00 from alaris. hope this helps