Thu Mar 17 2011

error code 104?

what does error code 104 refer to?

   RE: Mobile Radiography Report Post
Thu Mar 17 2011Reply from John Storch
John Storch
Error 104 Charge Fault Charging current feedback value was low during trickle charge. Was charging initiated immediately after the unit was just charged? The charger may saturate faster than the firmware can compensate in this case. Otherwise (1) Recalibrate Charger (2) Replace the Charger Board ~ John

Report Post
Thu Mar 17 2011Reply from ruforjc2
ruforjc2
You da man John! Thanks for the insight! Mike

Report Post
Thu Mar 17 2011Reply from AMX Solutions, Inc.
avatar placemark
I spoke with our lead engineer and he said Error Code 104 is a charger fault. Look at J11 and J5 connectors on the charger board for burnt pins, especially the J11. If they're burnt, replace the pins and the Jones plug. If you need further assistance, please call us a 866.630.2697 and ask for Dave Birmingham. I hope this helps!! Maggie O'Mahoney

Maggie O'Mahoney
AMX Solutions, Inc.
For Technical Support, call 866-630-2697
Report Post
Thu Mar 17 2011Reply from ruforjc2
ruforjc2
Thanks Maggie, you da gal!!! : )

Report Post