Sat Nov 17 2012

Hitachi EUB-525 service manual

Hi

i am looking for Hitachi Eub-525 service manual

 

thanks in advance

xray.technician@yahoo.com


   RE: Ultrasound Systems Report Post
Mon Nov 19 2012Reply from MedWrenchManager
MedWrenchManager

 

MedWrench is available to help people with specific questions and troubleshooting problems. Could you please be more specific with the issue that you are having with your instrument , and the MedWrench team will do their best to assist you in solving the issue. This will also help other community members who face similar challenges.

 

 

Report Post
Wed Jan 22 2014Reply from LeoUsam
avatar placemark

Dear Sir / Mrs.

I am trying to fix a Hitachi EUB-525. Can you help me with a problem of  the power supply, I do not have schematics for this unit and I identified the defective power supply but I am unable to fix it without a schematic or a hint. One of the 5V supply is not starting,  I measured and changed one of the capacitors that was off a bit, but still the unit does not work. Please help me with this problem.

Regards.


Report Post
Wed Jan 22 2014Reply from Gantry
Gantry

you could try contacting Providean. The sell the service manual.

(877)661-8224

http://www.providianmedical.com/service-and-repair/hitachi-ultrasound/


Report Post
Fri Jan 24 2014Reply from Abdul Qayum
avatar placemark

Sir,

I have not the schematics & service manual but i can repair it.I have worked on the power supply of Eub-525.

abdulqayum14@gmail.com


Report Post
Mon Jan 27 2014Reply from LeoUsam
avatar placemark

Dear Sirs,

The problem I had was due to faulty capacitors. Changed the bad ones and now all is ok.

Regards.


Report Post
Sun Nov 09 2014Reply from kingseed4u
avatar placemark

Please' I'm working on a Hitachi EUB-EUB-525 . This is the problem: 

When the system is power, after using it for two patients' then on the third patient, while on the measurement mode, then it will freeze and hang and will not respond again. If it is restarted, then after the first patient then on the second patient, then it will freeze and hang. If you shut down and restart, then the time for it to freeze and hang keep reducing. Can it be the software or the hard disk. Please I need help to solve this problem.


Report Post
Fri Feb 20 2015Reply from blister
avatar placemark

Did you clean top and bottom of pcb rack? Very common prolem.Overheating.


Report Post
Thu May 21 2015Reply from Manager
avatar placemark

There is no HDD in this system. Cleaning the system and checking every PCB in the Scan Converter may be a good idea. When you pull out the PCBs, check the chips if they are properly seated in their sockets. Due to temp. differences, they may work themselves out of the sockets they are in. Check all connectors as well at the same time.


Report Post
Fri Jun 12 2020Reply from Rosmary
avatar placemark

Hola tengo un equipo Hitachi 525 el problema que hoy en dia tengo es que cuando lo encendia este congelaba la imagen al cabo de 5 segundo y no me permitia acceder a los botones de la consola no podia medir. Lo apagaba lo encedia volvia por unos breves segundo hacer examen y volvia la falla. Luego  llego hasta el momento enciende el monitor no emite imagen. Ya se reviso pila fuente poder este no es. Podrian ustedes ayudarme  que mas o menos podria ser? Aparte necesito manual de servicio de este equipo hitachi 525 donde lo compro o me lo facilitan y tambien sotfware por favor .


Report Post
Mon Jun 15 2020Reply from Ben C
Ben C

 

Aunque los consejos sobre el software no se pueden compartir, una de estas compañías puede ayudarlo con su problema técnico:

Agiliti 

Conquest Imaging 

Exclusive Medical 

Innovatus 

Probo