Thu Dec 06 2012

System Failure 11

Anyone come across this System Failure 11?  I've tested all the test points on the boards and their good.  I also changed the thermistors and the ribon cable between the boards.


   RE: Infant Incubators & Warmers Report Post
Wed Nov 23 2022Reply from Mport
avatar placemark

Hello, we are seeing this system failure #11.  did you ever figure out what the issue was causing this error code?  Thanks!  Mark

 


Report Post
Wed Nov 23 2022Reply from Chris S
Chris S

Have you followed the troubleshooting steps in the Service Manual starting on page 149?


Official MedWrench Guru
Report Post
Wed Nov 23 2022Reply from Bill L.
avatar placemark

Check fuses F9 and F10, possibly replace the Power PCA board. 


Report Post