What about said failure "Transmit Overload" after booting of system ?

Thank You !


   RE: Ultrasound Systems Report Post
Mon Dec 10 2012Reply from GaryBiomed
GaryBiomed

The manual says something about a front end processor problem

You can get the service manual here

http://www3.gehealthcare.com/en/Support/Support_Documentation_Library

You should contact Trisonics for tech support

http://www.trisonics.com/

or

Unisysn

http://www.unisynmedical.com/


Report Post
Mon Dec 10 2012Reply from dferguson
dferguson

TD or FEPS in most cases


 
 

Report Post
Fri Jun 10 2016Reply from Paul B
avatar placemark

The following diagnostic test can be utilized to determine if a Time Delay (TD) board is associated with error condition 'Transmit Overload' on a GE LOGIQ 9 ultrasound system:

1. DISCONNECT ALL PROBES from the machine to prevent high voltage damage during the test.

2. Enter service mode using account ‘External Service’.

3. Navigate to ‘Diagnostics’ --> ‘Acquisition Diagnostics’ --> ‘High Voltage Fault Diag’

4. Verify all probes are disconnected from the machine and all shielding panels are installed before continuing.

5. Read and follow all manufacturer procedure instructions in the right side column before starting the test.

6. Test results will be displayed in the right side column. If a TD board is defective, a message similar to ‘Failed: TD Board [slot number],…..’ will be displayed.

7. If a TD board is at fault, the problem can be re-verified by doing the following:

i.    Shutdown the acquisition computer

ii.   Remove power from the system

iii.  Using electrostatic discharge prevention tools, remove and mark the suspect TD board with permanent ink to prevent mixing it up with other boards. Swap the TD board with another slot position that contains the same type of TD board (ie. 'TIME DELAY 4, 2260194-2C').

iv.   Reinstall the metal shielding plate over the card cage for safety during testing.

v.    Power up the system and repeat steps 1 through 5 above.

vi.   If reseating the board connector, by moving it, did not resolve the issue, the displayed test results will indicate the same TD board failed in the slot position it had been moved to.