what is a system fault 4.2 on the candela YAG ?


   RE: Aesthetic Laser Report Post
Thu Mar 07 2013Reply from CRC Biomedical Services
CRC Biomedical Services

This is a power supply fault. Please contact me if you need this corrected.

 

Thank you.


Mickey A. Couvertier, CBET, CLRT
President & Senior Technician
CRC Biomedical Services
http://www.crcbiomedical.com
ph. 202-321-9947  em. mickey@crcbiomedical.com

Report Post
Fri Dec 30 2016Reply from Debra721
avatar placemark

what does system fault 10.4 mean on agentle yag?

 


Report Post
Mon Jan 02 2017Reply from kelsey
avatar placemark

A fault 10.4 occurs when laser enters ready state and did not detect the fiber.

Make sure that the fiber connector is properly inserted.

Check the fiber switch, connector and wire harness, CPU I/O PCB.