No exposure, no error Optimus 50 generator

 

When depressing the two stage switch to the second position, the exposure never takes place. I hear the high speed rotor during prep and then pressing shoot, nothing. No error response. The ready light and high voltage indicators are showing green on the optimus console. Any suggestion would be appreciated as I am I volunteer in Tanzania and this very busy hospital has only a portable working at the moment.

 


   RE: R/F Systems Report Post
Mon Mar 25 2013Reply from eb66401
avatar placemark

 

I have acquired some additional information. It appears that the unit will expose correctly but only when using the above-table option. Under-table and vertical selection will not expose. This makes me think that there is a signal for the cartridge that is not making it back to the control section. Is anyone familiar with this problem?  Thanks.


Report Post
Mon Mar 25 2013Reply from MedWrenchManager
MedWrenchManager

It sounds like there is not a cassette present signal from under table Bucky releasing exposer.

Answer provided by Rick P., MIT


Report Post
Mon Mar 25 2013Reply from eb66401
avatar placemark

Thanks Rick. I'm not sure but I assume there is some kind of limit switch for the under table cassette, is that correct?


Report Post
Tue Mar 26 2013Reply from www.PI.solutions
www.PI.solutions

When you press the test button on the overhead does it tell you anything?



Jamison Ashby, MBA

Owner: Precision Imaging solutions

www.PI.solutions
Report Post
Tue Mar 26 2013Reply from MedWrenchManager
MedWrenchManager

Its most likely a sensing parameters depending on cassette size.

Answer provided by Rick P., MIT


Report Post
Tue Apr 02 2013Reply from eb66401
avatar placemark

Thank you for all the replies and good ideas. I have found that the limit switch for the cartridge in both the table and vertical Bucky is bad. I've repaired the Switches and all is fine. Thanks again.


Report Post
Thu Jan 09 2014Reply from tiralb123
avatar placemark

Hi,

 

I have the same problem, can I have your email?

my email address is tiralb123@gmail.com

 

regards


Report Post
Sun Feb 09 2014Reply from grAXI
avatar placemark

green donut on before attempted exposure? Should go out and come back on rotor up to speed, filament boost after that the only interlock is the grid!


Report Post