Wed Jun 05 2013

TritonDts Rope Issue

My rope is stuck inside the head of my unit. Can I fix this myself or should I call a technition?


   RE: Traction Table Report Post
Thu Jun 06 2013Reply from MedWrenchManager
MedWrenchManager

turn on the unit and try to release the rope using the unit fuction, if it does not, turn the unit off and call your Technical service provider, the unit may need calibration.

Repost from MedWrench User.


Report Post
Thu Jun 06 2013Reply from VMessina
avatar placemark

Thank you for your advise. I was given re-calibration instructions, however, the rope is still stuck in the head. I called a technician.


Report Post
Thu Jun 06 2013Reply from Sebastian
avatar placemark

Do not turn the unit on again !

You should wait for the technician, He/She should calibrate the unit using special tools and instruments, the problem could be the load cell, in the worst case the load-cell must be replaced. 