I have a CCT giving this error, is there a fix to this issue besides replacing the system board? I contacted Zoll and they instructed to try to clear the error and perform a self test (which i did both), no fix, so they informed me that the system board is defective. It's a $3828 board, is there another solution or is a board replacement the only option? Thank you for any help.
ICU error 314
You can send this unit in to Image Quest they repair defibrillators to component level at a very cost effective flat rate. You just have to send in the whole unit to them with accessories so that once they are done with the repair they can fully test and burn in the unit. They can be reached at 1-855-813-8100
Report Post
The Isolated power supply board says "IBP/TEMP/NIBP POWER" in small white print. The part number is on it also, look for "ASSY #9307-0349". The board is located at the bottom of the CCT and is square in the middle of the case. It has a ZIF connector on it that connects via ribbon cable up to the system board. The board it self is about the size of a standard playing card. Hope this help.
Report Post
Thanks Francis. I found it seconds after I posted. Unfortunately, it didn't fix my problem.
Report Post
Oh no! Well it could be the ribbon cable or the system board at fault then. I have also received bad replacement boards from Zoll in the past.
Report Post