Monocular readings in the Kerameter ARK 900

Our ARK 900 Refractor / Keratometer does´t change when the right eye has been measured, the left eye reading appears as a continued measurement of the right eye despite is set to automatic binocular measuring.


   RE: Ophthalmic Equipment Report Post
Tue Sep 03 2013Reply from GaryBiomed
GaryBiomed

You may want to contact Marco

http://www.marco.com/contact.html

1 (904) 642-9330


Report Post
Tue Sep 03 2013Reply from see2020
avatar placemark

Generally, there is a "finger" under the platform which articulates left and right to help with this determination. I have found, based upon my collegues recommendation, that when I pass a stiff piece of paper (or a single thickness of a medical chart is better) under the platform, I am able to cause the finger to rock left/right, and this 'fixes' the problem... Give it a try.. if not, I have an excellent repair person (not Marco/Nidek) who has worked with me for 15 years, on any of my 3 ARK-900 machines.... no hesitation to recommend him. Just shoot me an email (see2020@att.net).  DJMedford, MD