CX50 boots up with an error...

Dear Members,

The CX50 machine boots up completely till user area appears and then it gives "Call service...' message.

The application log shows "HV calibration failed"

Does anybody faced similar situation or is there any error/troubleshooting guide?

Any advice will be appreciated...

Artur 


   RE: Ultrasound Systems Report Post
Mon Jan 20 2014Reply from Jim Carr
Jim Carr

Howdy Artur,

At AUE, we have seen this caused by a defective Power board. It is also possible but less likely that a defective Channel board could cause that error.

Jim


Jim Carr
Director of Service & International Operations
AUS
Toll Free (USA, Puerto Rico and Canada): 866-620-2831
USA office: +1 918-628-2831
Email: Service@advancedultrasound.com and Sales@advancedultrasound.com
 

Report Post
Tue Jan 21 2014Reply from ArturCuriosity
avatar placemark

Hello, Jim!

Thanks for reply.

RST shows 2 failed tests: CBHVp current test and CBHVp ramp test. Other tests pass pretty fine.

Does it narrow down the list to Channel board?

Is there any possibility for software glitch or wrong Bios settings?

Thanks in advance!


Report Post
Tue Jan 21 2014Reply from ArturCuriosity
avatar placemark

Hello, Jim!

Thanks for reply.

RST shows 2 failed tests: CBHVp current test and CBHVp ramp test. Others pass pretty fine.

Does it narrow down the list to Channel board?

Is there any possibility that software glitch or wrong BIOS setting could lead to such results? Soft version is 2.0.3.

Thanks in advance!


Report Post
Tue Sep 13 2016Reply from mukapotap
avatar placemark

Hi ArturCuriosity, how you can get acces to RST? My email: mykytenkoevhen@gmail.com