HIGH SODIUM AND CHLORIDE RESULTS
Wed Jan 22 2014
RXL IMT ELECTROLYTES
Mon Feb 03 2014Reply from Irfanlab
Hi,
This is common issue, fallow below mention steps;
1. check the expiries of Standard A, B, Flush, Salt Bridge & Sample Diluent. If so change it.
2. verify the correct position of tubing connections.
3. Do the IMT bleeching and change the sensor.
4. Reply me if still problem persist.
Irfan
Lab Specialist
KAMC KSA.