RXL IMT ELECTROLYTES

HIGH SODIUM AND CHLORIDE RESULTS


   RE: Clinical Chemistry Analyzer Report Post
Mon Feb 03 2014Reply from Irfanlab
avatar placemark

Hi,

This is common issue, fallow below mention steps;

1. check the expiries of Standard A, B, Flush, Salt Bridge & Sample Diluent. If so change it.

2. verify the correct position of tubing connections.

3. Do the IMT bleeching and change the sensor.

4. Reply me if still problem persist.

Irfan

Lab Specialist

KAMC KSA.