Mon Mar 31 2014

SIEMENS SONOLINE ANTARES( VGA CARD REPLACING)

I HAVE SIEMENS SONOLINE ANTARES THAT JUST REPLACING NEW VGA CARD THEN THE SYSTEM NOT SHOW GOOD SIGNAL (SMALL SCREEN AT LEFT CORNER OF DISPLAY), THE SYSTEM SAID "PLEASE CHANGE THE DISPLAY SETTING OF YOUR SYSTEM"

ANY ONE KNOWS HOW TO LOG TO WINDOW TO CHANGE DISPLAY SETTING OR HAVE SERVICE MANUAL?


   RE: Ultrasound Systems Report Post
Wed Apr 02 2014Reply from domessj
avatar placemark

check the connector that goes from the CPU board to the video card that has not been disconnected, the resolution should take it alone

 


Report Post
Thu Mar 15 2018Reply from service.ultrasound.ct.mri@gmail.com
avatar placemark

Hi, if you want to go in Windows mode you need make next steps:

Press Setup menu (Presets), next "Service". In Service menu open field Printer Install and will open small  window. Press 2 times Backspace button, this will open My Computer window. Next you can open Control Pannel to change display settings.

 

If you will have more questions write on

service.ultrasound.ct.mri@gmail.com 


Report Post

This site uses cookies. By continuing to browse the site you are agreeing to our use of cookies. Please review our Privacy Policy for more details.