Tue Apr 08 2014

IE33 Video Boards

Does anyone have the part number for the video boards in an IE33 E Cart?

Thanks,

Terry


   RE: Ultrasound Systems Report Post
Tue Apr 08 2014Reply from BMET-KY
avatar placemark

The part number is 453561270341 but only Philips engineers can order it from Philips. They will tell you to order the UMB assembly for around 11000. I have had good luck from conquest imaging for around 1000 for the video card.


Report Post
Tue Apr 08 2014Reply from BrownR5
BrownR5

MSI NX7600GT is the video card if you can find someone that still has a reliable card. These video cards were discontinued from MSI so the trick would be to find an equivalent card that uses the same drivers. I lucked out and found one at Summit Imaging. That was three years ago so you may have to contact MSI and see if they have a compatible video card with the same drivers.


Report Post
Tue Apr 08 2014Reply from dferguson
dferguson

Which E cart are you in?  If you are in an early E cart then you would have 7600GT and the Later, which I believe was E.2, would have 8600GT.  They are not directly interchangeable.  If you have the 7600GT, the card should be part number 453561270341 or its' alternate 453561620181.  The 8600GT is 453561344971.


 
 

Report Post
Tue Apr 08 2014Reply from dferguson
dferguson

Which E cart are you in?  If you are in an early E cart then you would have 7600GT and the Later, which I believe was E.2, would have 8600GT.  They are not directly interchangeable.  If you have the 7600GT, the card should be part number 453561270341 or its' alternate 453561620181.  The 8600GT is 453561344971.

 If you need either one, call Conquest Imaging @ 866-900-9404.


 
 

Report Post
Wed Apr 09 2014Reply from Terryt
avatar placemark

This is the 7600GT board. Thanks for the replys.