Tue Apr 22 2014

pump continuously beeps with error (6,3) when turned on

What is causing this error?

Thank you for your help!

 


   RE: Infusion Pump Report Post
Thu Jun 19 2014Reply from BioMedovation
avatar placemark

Not seen that error message before. Image Quest repairs these units at a very cost effective flat rate. You can contact them at 1-855-813-8100


Report Post
Wed Jun 25 2014Reply from JSchwager
avatar placemark

It's most likely an AIL issue. Check your Air-in line sensor cables on the backside of the MEA board. Sometimes they get caught in the housing when they get put back together and the cables can get cut/shorted causing this error on power up, or the sensor itself (inside the AIL housing) could possibly be going bad.

Also, we repair/service and sell parts for Medsystems at Biomedix Medical. Feel free to call us for help or parts  610-268 -1444