STERIS - AMSCO 3080
by STERIS

STERIS AMSCO 3080 Forums

Updated  Oct 22 2021
1 Reply

This just happened today, I can tilt and perform all adjustment but the table will not auto level. I tried a different hand control ( pendant ) from another working bed and the same result. Thanks Dwight

RE: STERIS - AMSCO 3080
Added  Sep 22 2021
0 Replies

Hi all, I hope that you can help me on an Amsco 3080RL surgical table. This table is leaking oil and the only thing that I found on the table was a piece of tubing (that goes from the foot pump to the electrical pump) has split and the table is low on oil. My question is: what type of oil do I use and is that possibly where it was leaking? Thank you, Mike

RE: STERIS - AMSCO 3080
5 Replies

What is the use of this tubing? It connects the manual pedal pump to the electric hydraulic pump

RE: STERIS - AMSCO 3080
Updated  Nov 6 2019
1 Reply

Hi everyone, we have an electric operating table (steris Amsco 3080) which is having a few problems, I dont know whether this model is able to be operated manually, ie is it possible to override the electric function of the table?

RE: STERIS - AMSCO 3080
Updated  Mar 16 2019
2 Replies

Need a copy of the operator Manual.

RE: STERIS - AMSCO 3080
Updated  Sep 1 2016
3 Replies

i have problems with trendelenburgh movement. Valve 13, 10, 8 and 7 work properly. What else can i do?? 

RE: STERIS - AMSCO 3080
2 Replies

Need procedure to replace gears on the kidney rest shaft

RE: STERIS - AMSCO 3080
Updated  Jul 28 2016
3 Replies

I'm looking for the hydraulics diagram or service manual for a old version of the 3080 with Parker(blue aluminum hydraulic assemblies0 not the newer Swiss made Bieri hydraulics. Thanks.  ernestoquintero747@gmail.com

RE: STERIS - AMSCO 3080
Updated  Oct 29 2015
3 Replies

I was replacing the main batteries on a 3080 table today and noticed a leak. After removing the batteries is was apparent the leak is from a corroded hose right by the battery, must of been slight rubbing the hose.

After contacting Steris for help identifying the part, I was told this particular batch of 3080s with Bruning Hydraulics system is no longer supported.

I am needing to replace is the hose that operates the head end floor lock at the end of the table underneath the heavy counter weight.

Anyone have a source for this part ?

RE: STERIS - AMSCO 3080
Updated  Feb 16 2015
2 Replies

Will power up the pendent but it is not working to maneuver the bed. It flashes that the bed is unlocked.

RE: STERIS - AMSCO 3080