I just turn ON the system and remains at TESTING MEMORY and do not go to main screen. I replaced TIMEKEEPER BATTERY, but It did not doluion this fail. Any recommendation are welcome on how to solve this problem.
RE: STERIS - Amsco Century V116I AM GETTING A LEAK RATE OF 4.0. I REBUILT THE STEAM MODULE, ALSO THE EXHAUST MODULE. NEW GASKET, CHECK VALVES ON THE HEAT EXCHANGER. NEW HEAT EXCHANGER. COULD I BE GETTING THE LEAK THROUGH THE FLOAT SWITCH INSIDE THE CHAMBER?
RE: STERIS - Amsco Century V116Any ideas? I was doing a pm and had issues with reaching pressure and temp during cycles. Found a hole in jacket steam trap and a warn busing where the jacket rtd is inserted into plumbing causing a leak. Replaced both and calibrated temp/pressure (x2). Checked my PRV and the feed line a the top for blockage. Checked S9. Not sure what is going on.