Leica Microsystems - RM2255
by Leica Microsystems

Leica Microsystems RM2255 Forums

5 Replies

I have a Leica RM2255 that has a problem. The course feed forward and backward, on both fast and slow there is no movement. I can hear ticking sound but there isn't any movement. Is there a coupling inside between motor and lead screw that may be loose or failing. Any thoughts would be appreciated. 


Thanks

RE: Leica Microsystems - RM2255
Updated  Jun 29 2023
4 Replies

Hello,

I'm having a hec of a time trying to find a torsion spring/leg spring for the hand wheel lock, Part# 0502 38987 per the Leica manual. I've tried to modify some springs that I had in hand using pliers and clamps and that didn't go well at all. I've reached out to Leica and "crickets". Any insight in this matter will be greatly appreciated. 

RE: Leica Microsystems - RM2255
2 Replies

Hello, I am having issues where auto mode is not working...I have checked the PCB and noticed that R63 is damaged/cracked in the middle. Will that be the source of the problem? I have checked the test points as well in TP2 and everything seems to be working. Any help will be appreciated! Thank you in advance!

RE: Leica Microsystems - RM2255
Updated  Mar 19 2020
1 Reply

How to Reset Service Message

RE: Leica Microsystems - RM2255
1 Reply

hi I have one old type of auto microtome with RM2155model for general service , its dot point of display in the right side of display still turn on all the time as I check in its user manual that is service counter indicator ,so how I can rest it ?

RE: Leica Microsystems - RM2255
Updated  Dec 2 2019
3 Replies

Issue seams to be when rotating the handles wheel mechanism squeaks. Any recommendation of whaer and should I lubricate or grease and with what thanks.

RE: Leica Microsystems - RM2255
Updated  Sep 5 2019
2 Replies

I have a unit that will move forward and back with the control but in the automatic during operation it wont. Looking at it tomorrow any quick ideas to check. Thanks

RE: Leica Microsystems - RM2255
Updated  May 3 2019
6 Replies

We have a microtome that you can push start and stop for about 5-10 times than it freeze&.the buttons LED i Lightning but nothing happends when you push the buttons. if you turn it off and on again it works for a short time and then there is the same problem. The controlpanel + cabel i new and the motor controlboard is new. Any ideas what cause this problem ?

RE: Leica Microsystems - RM2255
Updated  Feb 7 2019
2 Replies

Since I replaced the controller board, the specimen head stops in the down position rather then the top position in auto mode. How can this be corrected?

RE: Leica Microsystems - RM2255
6 Replies

hi I have a RM 2255 with destroyed Control electronics PCB its Control electronics PCB is different with service manual that are in the Documents tab !!!! I see that is new type with SMD Electronic components... so any body have manual or Wiring diagram of new type and any body have the Control electronics PCB for sale ?