Updated Jul 7 2021
3 RepliesI have replaced the garments and the tubing, and I have replaced the valve manifold. Equipment still won't pass T6. I get the low pressure error. Any ideas? RE: Kendall - SCD EXPRESS
Updated Sep 8 2020
Updated Aug 20 2018
5 RepliesSCD EXPRESS SYSTEM ERROR ON INITIAL POWER UP.
ERROR OCCURS REGARDLESS OF BATTERY INSTALLATION.
BATTERY SHOWS NO CHARGE
BATTERY WILL NOT TAKE A CHARGE FROM EXTERNAL CHARGER
NO SPECIFIC FAULT CODE IS CALLED OUT BY UNIT.
24Vdc OUT OF J4 (J1 PER MANUAL) ON PS-BOARD TO CPU-BOARD
0V OUT OF J6 ON PS-BOARD TO 6Vdc COOLING FAN
INDUCTOR L3 FEEDS BOTH J4 (J1 PER MANUAL) AND J6
295 OHMS ACROSS COMPRESSOR
4.6 MOHMS ACROSS COOLING FAN
50 OHMS ACROSS (6) RELIEF SOLENOIDS
I SEE NO OBVIOUS SIGNS OF DAMAGE ON THE SECONDARY BOARD.
I'VE REACHED OUT TO J2C WITH MINIMAL RESULTS.
HELP AND FEEDBACK WOULD BE APPRECIATED..
SEMRAD RE: Kendall - SCD EXPRESS
Added Oct 29 2016
0 Replies
Had this Kendall SCD Expressunit since 2008. Wife started having problems with her legs again. Got the unit out of storage plugged unit in, turned unit on, pump stokes 2 times and I get a beep and notification "System Error Service Required". Where can I send this unit for service.
RE: Kendall - SCD EXPRESSUpdated Oct 30 2015
Updated Apr 7 2015
1 Reply
Does anyone know distributers to get Kendall SCD parts from? Thanks in advance.
RE: Kendall - SCD EXPRESSUpdated Aug 1 2014
1 Reply
Turn power on, compresor starts initially, then unit shuts off and will not power up unless remove battery for 5 seconds.
RE: Kendall - SCD EXPRESSUpdated Jun 5 2014
Updated Apr 28 2014
Added Apr 8 2014
0 Replies
What is a good way to remove the "Bed Hooks" from the SCD?
There must be a smooth way. I know about pressing the bars together, but still even this is a struggle at times.
Any advice would be greatly appreciated