GE HealthCare - Precision 500 D
GE HealthCare Precision 500 D Forums
Add R/F Systems to My Bench27 Results
Added Dec 19 2022
0 RepliesGenerator error 3601163 is being triggered everytime I try to use the large focal spot. I already measured resistance and continuity directly into the tube and also in the generator side to make sure cables are ok as well. They are just fine.
I also did what the manual says, I replaced the heater board and it is still the same.
Does anyone have a recommendation? RE: GE HealthCare - Precision 500 D
Added Aug 25 2022
0 RepliesHave an isse with table angulation at zero deg reference, not consistant, motr, pot, board and MRAS controller all have been changed out, loaded software and backups. any ideas?
Equipment: GE Healthcare - Precision 500 D
RE: GE HealthCare - Precision 500 DAdded Aug 25 2022
0 RepliesI have an issue with A GE prcision 500 table, the table has zero references that change (ex. 0 deg ref to start, if i power th room down or reset it may be at -5 deg or another number there is no consistancy), motor pot,board, MARS system controller have all been replaced, all software reloaded and table calibrated multiple times. Any ideas? RE: GE HealthCare - Precision 500 D
Added Jul 15 2021
0 RepliesHi everybody.
My device is a GE P500D and when it works there is no overload warning sound. This is going to be one of the big problems. I want to install a special sound system for it is it possible or not. Can someone who is a technician give me some suggestions? RE: GE HealthCare - Precision 500 D
Updated Apr 23 2021
2 RepliesProblem: Getting optics errors...codes not available at this moment. Density filter is too slow to move into position and causes the errors. I have lubed it to get by for a few years, but it is getting worse...same errors on three P500D units we own. Apparently this happens as the units age. Anyone know of a source for an equivalent density filter actuator? Uses a 5-pin servo actuator, like is used by hobbyists in Remote Control miniature cars, boats, helicopters, planes, etc. I have removed the actuator/density filter assembly, but dont see the manufacturer label. It could be under the glue that holds the small circuit board onto the actuator. Hate to spend tens of thousands of dollars when part needed should cost less than $200. RE: GE HealthCare - Precision 500 D
Updated Dec 8 2020
Added Apr 1 2020
0 Repliesissue moving the bucky easily, thinking the power assist might be bad. RE: GE HealthCare - Precision 500 D
Added Jul 31 2018
0 RepliesWe are currently upgrading a GE P500D from CR to digital and need to disable the bucky movement. Can this be done in configuration or with jumpers/dip switches. Thanks,, RE: GE HealthCare - Precision 500 D
Updated Apr 8 2017
12 Repliesgood afternoon everyone,does anyone can help me with follow problem,I am facing a problem with Precision 500D table angulation.it doesnt tilt and I can see the K2 relay energized but not K1,could it be the cause of problem? RE: GE HealthCare - Precision 500 D
Added Oct 10 2016