Draeger - Permox Silent Care
Draeger Permox Silent Care Forums
Hello,
I have a Draeger Permox Silent Care with the following symptoms:-
- If switched on with airflow restricted (4Lpm for example), then it triggers an immediate alarm mode and shuts down the compressor.
- If switched on with airflow unrestricted (regulator fully opened) it runs just fine and produces oxygen. If left to run, after 5 minutes you get a 3-beep alarm, but it continues to run.
- One it is running unrestricted, if the airflow is restricted (to 4Lpm for example), then after a while it will go into a single continuous alarm and shut down the compressor.
- I have checked all the pipes, no kinks or snags found, and with the regulator fully open it easily exceeds the flow measurable by the flow meter.
- Filters are all clear.
- Compressor runs nicely
Can anybody please advise on what may be wrong?
I notice there is an RS232 port and also a header pin block on the PCB. Can I use these to aid diagnostics?
Is there ANY technical or service documentation available apart from the standard use manual?
I am located in Hampshire U.K.. Are there any service places nearby?
Many thanks for any help or advice.
RE: Draeger - Permox Silent CareWe have the Permox SilentCare, but not used it for a long time. In light of the COVID-19 we wish to have the machine operational, but when switching on the red light blinks and every few seconds there is a sounds signal. We don't have the manual available so we are seeking a link to that, or an advise on what to do to have the machine ready.
Thank you.
RE: Draeger - Permox Silent Careservice manual needed fordrager permox silentcare oxygen concentrator type G4h80 serial NO SRPB-0055
Hi
is there any one who can help me with the service manual of the drager permox silentcare oxygen concentrator type G4h80 serial NO SRPB-0055 any help will be appreciated
RE: Draeger - Permox Silent Care