Acuson Sequoia Ultrasound System Third Party Service Tip from Conquest Imaging
Here is a tip for the Acuson Sequoia Ultrasound System.
Wed Oct 23 2013
Sequoia Third Party Service Tip
When accessing an Acuson Sequoia card cage, be aware that there is a security system in place that will disable the system unless certain conditions exist:
- That the “Third Party Service” option is labeled “Permanent” in the “Feature Control” menu
- The system is at or above revision 7.2 software and is a Phase 3 hardware system. The easiest way to identify a Phase 3 system is the absence of an RJ45 (network) connector in the top-left area inside the rear panel where all the input/output connectors reside (all systems have an RJ45 connector on the exterior just below the handle in the rear). If you see an RJ45 connector inside, the system is a Phase 1.
- If these either of these conditions are not met, you must contact Siemens for the “Third Party Service” option code. No special access is needed to install the code/option so please contact Conquest Imaging for assistance.