Class 1 Recall: Heart Sync, Multi-function Defibrillation Electrodes Will Not Work with Philips FR3 and FRx AEDs
A recall has been issued for Heart Sync, Multi-function Defibrillation Electrodes because they will not work with Philips FR3 and FRx AEDs. Philips made changes to the connector design of their FR3 and FRx AEDs, and because of these changes, the
Thu Jan 08 2015
This recall was updated on January 8, 2015.
Recall Class: Class I
Date Recall Initiated: November 11, 2014
Devices:
All lot codes of the following AED electrodes distributed between October 26, 2011 and November 26, 2014
Adult Radiotransparent Electrode
- Catalog Number C100-PHILIPS
Adult/Child Radiotransparent Electrode
- Catalog Number C100AC-PHILIPS
- Catalog Number T100LO-PHILIPS
Radiotranslucent Electrode
- Catalog Number T100-PHILIPS
Adult/Child Radiotranslucent Electrode
- Catalog Number T100AC-PHILIPS
NOTE: These electrodes will work with other Philips AEDs that accept plug style connectors.
Use: Electrodes are connected to an automatic external defibrillator (AED), which analyzes the heart rhythm in cardiac arrest patients and delivers an electrical shock to restore normal heart rhythm. The primary users of AEDs are first responders and hospital health care providers.
Recalling Firm
Heart Sync, Inc.
5643 Plymouth Road
Ann Arbor, Michigan 48105-9586
Reason for Recall:
These defibrillation electrodes have been recalled due to a connector compatibility issue with Philips FR3 and FRx Defibrillators.
The FRx AED requires electrode pads be connected to the device before it is used. The AED will make a continuous alarm chirp to alert the user that the correct pads are not connected.
The FR3 does not require electrode pads to be pre-connected. Users will not know that the pads do not work until they try to use the AED. This may result in a delay in delivering the electrical therapy needed to revive a patient.
Delay in therapy could result in serious injury or death.