Class 1 Recall: CareFusion Recalls AVEA Ventilator Due to an Electrical Issue Which May Cause an Unexpected Shutdown
CareFusion is recalling the AVEA Ventilator because of a faulty fuse on the ventilators’ alarm board, which may cause the ventilator to unexpectedly shut down.
Fri Jul 29 2016
The FDA has identified this as a Class I recall, the most serious type of recall. Use of these devices may cause serious injuries or death.
Recalled Product:
Product Description
Product Numbers
AVEA Standard with Compressor ventilator-refurbished
R17312-xx (xx = 0-14)
AVEA Comprehensive ventilator
17310-xx (xx = 0 -14)
AVEA Standard ventilator
17311-xx (xx = 0 -14)
17312-xx (x=0 -14)
AVEA Comprehensive ventilator-refurbished
R17310-xx (xx = 0-14)
AVEA Standard ventilator-refurbished
R17311-xx (xx = 0-14)
R17312-xx (xx= 0-14)
TCA Board
16542A
GDE- 1st Generation
16222-001-99
GDE
16650A
GDE-refurbished
R16650A
AVEA GDE/UIM upgrade kit
12283-PMN
  • Manufacturing Dates: November 13, 2016 to January 4, 2016
  • Distribution Dates: to December 16, 2016 to February 15, 2016
  • Devices Recalled in the U.S. 501 units distributed nationwide
Device Use
The AVEA ventilator is intended for continuous breathing support for infants, children and adults. The ventilator is only used in hospitals and other health care facilities.
Reason for Recall
CareFusion is recalling the AVEA Ventilator because of a faulty fuse on the ventilators’ alarm board, which may cause the ventilator to unexpectedly shut down. If the ventilator shuts down, a patient may not receive necessary oxygen. The use of affected product may cause serious adverse health consequences, including death.
Who May be Affected
  • Health care providers using AVEA ventilators
  • All patient groups who may be using these ventilator systems for breathing support
What to Do
On May 17, 2016, CareFusion sent a "Field Safety Notice" letter to affected health care facilities. The letter instructed health care professionals and consumers to:
  • Complete and return the acknowledgement and receipt form
  • After the response card is returned, CareFusion will schedule onsite remediation for affected devices
  • Identify and remove affected ventilators as instructed in CareFusion’s Field Safety Notice
  • Report adverse events or quality problems experienced with use of the product to the FDA through:
Contact Information
Health care professionals and consumers with questions are instructed to contact CareFusion at 888-526-6018 orsupportcenter@carefusion.com with any questions related to this recall.
Date Recall Initiated:
May 17, 2016