Medical Equipment Alerts and Notices
MedWrench alerts and notices keep you informed and up-to-date on important safety information related to medical equipment.
Fri Jan 04 2013
Recent Medical Equipment Alerts and Notices
Class 1 Recall: GE Healthcare, LLC, Giraffe and Panda T-Piece Resuscitation Systems, and the Giraffe and Panda Bag and Mask Resuscitation Systems
The oxygen and air wall inlet fittings and/or labels on the back panel of the resuscitation systems may have been reversed during assembly. As a result, this could potentially reverse the air/oxygen concentrations; for example, a setting of 100% oxygen could have an output of 21% oxygen and vice versa. The settings of the blender knob would no longer be accurate.
DownloadClass 1 Recall: HeartSine Samaritan Public Access Defibrillator 300/300P
The firm issued a letter to customers on Sept. 11, 2012 describing the problem and provided instructions for correction. Certain Samaritan® 300/300P PAD devices have been found to intermittently turn on and off, which may eventually deplete the battery. In addition and separately, certain Samaritan 300/300P PAD devices containing early versions of the battery management software may misinterpret a temporary drop in battery voltage as signaling a low battery resulting in the device turning itself off. A device experiencing either condition could be unable to deliver therapy during a cardiac event.
DownloadClass 1 Recall: Mindray A3 and A5 Anesthesia Delivery System
The affected anesthesia delivery systems may have a gasket leak which could cause an interruption of or inadequate patient anesthesia and ventilation, temporary or permanent patient injury, or death.
DownloadClass I Recall: Baxter Healthcare Corporation, Automix Automated Nutrition Compounder Systems
Baxter Healthcare Corporation (Baxter) has initiated a voluntary recall of its Automix Compounder (Automix) because of incorrect key press responses, caused by fluid entry into device keypads, and intermittent electrical failures. Fluids, such as water, cleaning solutions, and nutrition source solutions, may enter into the keypad of the Automix control module and may cause the Automix to generate an incorrect device response to an Automix operator's key press. The intermittent electrical failures may cause the motors on the Automix to pump nutrition solution when not programmed to do so, or may cause the Automix to stop compounding before it has finished appropriately mixing the Total Parenteral Nutrition (TPN) Solution. Causes for the intermittent electrical failures have not been determined.
DownloadClass 1 Recall: CareFusion Alaris PC Unit, Model 8015
A component on the PC unit power supply board is causing an error code (120.4630): "System Error" or "Missing Battery Error" to occur. The error code is accompanied by both an audible alarm and a visual error message on the PC unit screen. If the error code occurs at start-up, the Alaris infusion pump cannot be programmed and this may cause a delay in patient therapy. If the error code occurs during infusion, the health care provider cannot make programming changes to current infusions, and this may result in serious injury and/or death to the patient.
DownloadClass 1 Recall: I-Flow ON-Q Pump with ONDEMAND Bolus Button
The firm determined that the ON-Q pumps with ONDEMAND bolus button may not lock in the down position when depressed and/or the orange bolus refill indicator may stay in the lowest most position. When this occurs, the patient may receive continuous infusion at a rate greater than expected. As a result, this product may cause serious adverse health consequences, including death.
DownloadClass 1 Recall: Respironics, Inc., Trilogy 100, 200, and 202 Ventilators
The Trilogy 100, 200, and 202 ventilators are being recalled because power supply components may be defective in some cases. If a ventilator’s power supply fails, the ventilator may unexpectedly stop functioning, thereby stopping ventilation. The ventilator alarm may also fail to sound. Ventilator power supply failures can result in serious adverse health consequences, including death.
DownloadClass 1 Recall Expansion: B. Braun Infusomat Space Infusion System
This is an expansion of a previous recall, initiated by B. Braun in January, 2012, due to the potential for breakage of the anti free flow clip catch located inside the infusion pump door. Breakage may occur when the IV set anti free flow clip catch is inserted improperly into the pump and the pump door is forced closed. Misloading of the anti free flow clip catch may create the potential for free flow of medication. Free flow, especially of narrow therapeutic range drugs, can cause life-threatening effects and injuries.
DownloadClass 2 Recall: GE Senographe 2000D Full Field Digital Mammography System
GE Healthcare has become aware of inaccurate measurements on GE magnified images provided by Senographe 2000D Mammography systems when reviewed on IHE compliant workstations. This issue may impact patient safety.
DownloadClass 1 Recall: Smiths Medical Medfusion Model 4000 Syringe Infusion Pump
System errors can cause the Medfusion 4000 infusion pump to go into an alarm condition and stop running, which can result in a delay or an interruption of patient therapy. For patients receiving critical therapy, such as vasopressors, a delay or interruption of therapy could result in serious injury and/or death. According to the firm, there have been no reports of patient injury or death due to this problem.
DownloadClass I Recall: Cardiac Science Powerheart, CardioVive, CardioLife; GE Responder and Responder Pro; and Nihon-Kohden Automated External Defibrillators (AEDs)
The affected AEDs contain a component that may fail unexpectedly due to a defect. If the component were to fail during a rescue attempt, the AED may not deliver defibrillation therapy causing serious adverse health consequences, including death. The unit’s self test may not detect the failure or impending failure of the component.
DownloadClass I Recall: CareFusion EnVe Ventilators
CareFusion has identified potential defects associated with the EnVe Ventilator model 19250-001 that can interrupt ventilation to the patient.
DownloadClass I Recall: Nemschoff Chairs, Inc., Perinatal Pediatric Hospital Bed (Bassinet)
Nemschoff Chairs, Inc. initiated a recall to correct issues with the wheels (casters), door hinges and drawer slides on the cart’s frame and the acrylic bassinet tub. Also, the manufacturer added labeling to the products regarding drawer weight limits, and specific instructions for cleaning the bassinet tub. This product may cause serious adverse health consequences, including death.
DownloadClass I Recall: CareFusion NicoletOne Software Used with CareFusion Cortical Stimulator Control Unit
CareFusion is recalling these affected devices and software for two reasons:
- The device’s software incorrectly indicates stimulation is delivered to a different electrode than the one selected
- A short circuit may develop between the cortical stimulator control unit and the stimulus switching unit amplifier
Class I Recall: Flight Medical V24-00400-29 Cable for Newport HT50 Ventilator
The recalled cable may cause an electrical shortage that can cause Newport HT50 ventilators manufactured prior to 2007 to shutdown unexpectedly.
DownloadExpanded Class 1 Recall:SIGMA Spectrum Infusion Pump Model 35700
Based on additional analyses since the initial recall, SIGMA expanded their recall to include additional affected units manufactured from January 18, 2005 through November 1, 2010, with the exception as noted above. These units may fail suddenly causing inaccurate flow conditions during use, ranging from back flow to over-infusion, including free flow. The pump does not issue an alarm when this occurs. These conditions could result in serious injury or death.
DownloadClass 1 Recall - GEM Premier 4000 PAK Cartridges for Use on the GEM Premier 4000 System
Potassium test results on the GEM Premier 4000 are too low when compared to a reference analyzer; with biases exceeding the allowable error claim of plus or minus 0.5 mmol/L by as much as 2.0 mmol/L. Use of this product may lead to inappropriate patient treatment and may cause serious adverse health consequences, including death.
DownloadSteris System 1 Processor: Extended Transition Period
Steris System 1 Processor: 6 Month Extension for Healthcare Facilities To Transition to Legally-Marketed Alternatives. FDA has extended the transition period to February 2, 2012. All healthcare facilities must be transitioned from Steris Corporation’s modified System 1 processor (SS1) to legally-marketed alternative devices by that time.
DownloadClass 2 Recall: Vertical Brake Hubs of the Computed Tomography X-Ray Systems
Philips Medical Systems has decided to recall the patient support vertical brake hub in all its Brilliance systems due to the collapse in the vertical direction. The short key connecting the vertical brake hub to the motor shaft can disengage due to the lack of thread locker on set screws.
DownloadRecall Class 1: Defibtech LLC, Lifeline and ReviveR AEDs
AEDs using software version 2.004 or earlier may cause the device to cancel shock during the charging process. Failure to provide appropriate therapy may result in failure to resuscitate the patient.
DownloadClass 1 Recall ACCU-CHEK FlexLink Plus
There is the potential to kink or bend the cannula when inserting the ACCU-CHEK FlexLink Plus infusion set. This can lead to under delivery and elevation of blood glucose levels.
DownloadClass 2 Recall: Precendence
Philips Healthcare Nuclear Medicine determined that the detector arm assembly could be compromised.
DownloadClass 2 Recall: Precedence
DownloadClass 2 Recall: SKYlight
DownloadClass 2 Recall: BuckyDiagnost with Optimus Generator
Affected systems were determined not to prevent run-on of a selected technique factor if the technique selector and the exposure control were activated simultaneously.
DownloadClass 2 Recall: AdvantageSim MD
GE Healthcare has become aware that inconsistent DRR (Digitally Reconstructed Radiographic) visualization may occur when the blended DRR feature or the Isocenter Move mode feature is used in combination with specified other functions of the AdvantageSim application on Advantage Workstation. These inconsistencies may affect patient safety. Other AdvantageSim functions are not affected.
DownloadClass 2 Recall BrightSpeed Excel/Edge/Elite Select Computed Tomography
Failure to properly document the CTDI in the Technical Reference or User Manual.
DownloadClass 2 Recall: GE Healthcare Discovery MR750 3.0T
Class 2 Recall: GE Healthcare Discovery MR450 1.5T
Class 2 Recall: GE Healthcare, Optima MR450w 1.5T MR systems
Class 2 Recall Toshiba AQUILION ONE
Incorrect Dose Calculation. The software that performs the pre-scan dose calculation uses an incorrect parameter for an exam Plan using a tube voltage of 80kV or 100kV.
DownloadHiSpeed X/i Computed Tomography operator consoles with the Model Number 2153675 and 2247802
A defect with the hardware and software displaying the incorrect dose data.
DownloadClass 1 Recall: Medtronic SynchroMed II and SynchroMed EL Implantable Infusion Pump and Refill Kits
Pocket fills (the unintended injection of drugs or fluids into the patient’s subcutaneous tissue at the pump pocket site instead of the pump) may result in patient harm, serious injury, and/or death due to drug overdose or underdose.
DownloadClass 2 Recall: Varian brand C-Series Ximatron Radiotherapy Treatment Stimulator, Model H72
There is a potential for failure of a lead screw drive system or image intensifier lead screw drive assembly, if normal maintenance inspections are not followed correctly.
DownloadClass 2 Recall: Varian Medical Systems 4D Integrated Treatment Console
Software anomaly may result in a design deficiency causing Plan Edit Settings to malfunction under specific circumstances, where the user selects to use a Cone Beam CT set up field.
DownloadClass 2 Recall: Varian Medical Systems Floorstand. An accessory of the Optical Guidance Platform
Potential patient impact -- limit switches for Floorstand may be inadequate for some models and allow for potential impact.
DownloadClass 2 Recall: Varian brand C-series Clinac, Trilogy and Novalis Tx, Software Versions 6.X and 7.X
The Auto goto or Auto setup functions in C-Series version 7 software ignore the couch angle and exceed motion zone (unless a tolerance is defined by the user); it has the potential for collision with the patient on the couch.
DownloadClass 3 Recall: Varian Medical Systems FastPlan software, Versions 5.5 and 5.5.1, Model Number HZ3
Due to software Issue, an error occurs when performing cone planning using MRI, which may result in the patient receiving a higher radiation dose than expected, potentially to healthy tissue.
DownloadClass 2 Recall: Varian brand Trilogy System with RapidArc
The system may deliver an inaccurate dose that does not match the treatment plan.
DownloadClass 2 Recall: GE BrightSpeed Excel/Edge/Elite Select / Elite CT Scanner System, models 2377708-560, 5377708-540, 5377708-560.
GE has become aware of a potential set of circumstances that could cause x-ray continuation during an unexpected table stop on certain LightSpeed VCT and BrightSpeed systems.
DownloadUrgent Device Correction Notice, Philips OmniDiagnost Eleva
Philips OmniDiagnost Eleva Failure Causing Generator to Stop Working. A short circuit may occur due to a faulty wire connection in the Supervisor Box Connector in the Generator. In exceptional cases if this occurs the generator stops working.
DownloadUrgent Field Safety Notice, GE Healthcare OEC 9800 C-Arm and 9900 C-Arm
Field Safety Corrective Action, HemoCue Glucose 201DM Software
HemoCue Glucose 201DM Software May Have Incorrect Configuation Settings, Error Codes
DownloadImportant Medical Device Correction: Medtronic 5388
Medtronic 5388 External Temporary Pulse Generator May Be Unable to Power Up or May Power Down Unexpectedly.
DownloadImportant Safety Information: PhysioControl LifePak
Important safety and disclosure information for professional users and consumers of the PhysioControl LifePak line of Defibrillators and AED's.
DownloadUrgent field Safety Notice, Leica Peloris Rapid Tissue Processor
Earth Leakage in a Single Fault Condition May Exceed the Allowable Current Limit. Recent tests have identified that under certain conditions the earth leakage, in a single fault condition, can be up to 6.8 mA which is above the allowable limit of 3.5mA.
DownloadUrgent Medical Device Correction, Terumo Advanced Perfusion System 1
Terumo Cardiovascular Systems (Terumo CVS) has received reports of premature failure of the oxygen sensor in the electronic gas blender of the Terumo Advanced Perfusion System 1.
DownloadDevice Recalls: Toshiba Aquilion ONE CT System TSX-301 A/2D
Electrocardiograph-Gated Dynamic Volume Prospective Scan May Cease Unexpectedly
DownloadMedical Device Correction, Philips IntelliVue G1 and G5
Critical Product Information - Cincinatti Sub Zero
Misusing Localized Heat Therapy for ECMO.
DownloadClass 1 Recall: Braun, Outlook 400ES Safety Infusion System, Model Number 621-400ES
Infusion systems upgraded with the Motorola compact flash hardware and supporting software when used in a network environment that utilizes Temporal Key Integrity Protocol (TKIP) authentication can potentially induce a memory leak that can cause the Management Processor to become non-responsive.
DownloadMedical & Radiation Emitting Device Recalls - TomoTherapy Hi-Art Treatment System
In some cases, the patient's diagnostic CT image is narrower than the Hi-Art radiotherapy couch image.
DownloadMedical & Radiation Emitting Device Recalls: Philips-Easy Diagnost Eleva
Potential for bright artifacts to appear on the patient image from a previous exposure.
DownloadUrgent Field Safety Notice, Beckman Coultre AC•T 5diff Hematology Analyzers
Valves May Fail Prematurely, Potentially Leading to Erroneous Results, Increased Instrument Errors, and/or Fluid Leaks
DownloadErrata Sheet for Package Insert
Package Insert May Contain Incorrect Acceptable Performance Range.
DownloadFDA Recall: Symbiq One and Two-Channel Infusers
Potential for unrestricted flow if cassette is removed before the cassette carriage fully opens and the slide/roller clamp is not closed.
DownloadFDA Recall: AngioSculpt EX PTCA
The PTCA catheters may become separated during use in which fragments of the catheter may become lodged in coronary arteries. This may result in serious injuries, including death.
DownloadGLOBAL RECALL OF MEDFUSION® SYRINGE INFUSION PUMPS
Class 2 Recall Innova 2100 IQ
Potential pump failure associated with the Coolix 4000 chiller of the Innova single plane and biplane systems that may stop the chiller from working. If a chiller failure occurs, the user will be informed with a message on the n-room monitor stating that three minutes are left before the X-ray is inhibited.
DownloadClass 2 Recall: Innova 4100 / 4100 IQ
DownloadClass 2 Recall: Innova 3100 / 3100 IQ
DownloadClass 2 Recall: Innova 2121-IQ / 3131-IQ
DownloadClass 2 Recall: STERRAD 200 Sterilizer
Oil mist filter fails allowing oil mist to be emitted into the vicinity of the sterilizer, which may result in the release of a mist, "haze," or "smoke."
DownloadClass 2 Recall: Alaris® Pump Module (a.k.a. Medley™ Pump Module)
The reason for this recall is that the units may contain misassembled occluder springs (bent, broken, nested or missing). These conditions have occurred due to misassembly during manufacturing.
Download