Class 1 Recall: Roche Diagnostics Recalls CoaguChek XS PT Test Strips Due to Inaccurate INR Test Results
Wed Nov 07 2018
Recalled Product:
- Name and Version: CoaguChek XS PT Test Strips
- Lot numbers: 28124111, 28124121, 28631911, 28631921, 28631924, 28632021, 28632213, 28632312, 28632412, 29415113, 29415123, 29494221, 29494312, 29494613, 29494711, 29778721, 29779012, 29779213, 29779214, 30497213, 30497311, 30497413, 30497423, 30497515, 31404314, 31404821, 32264116, 32264212, 32264316, 32264317, 32264411, 32264421, 33045913, 33046011, 33046113, 33046312, 33046314, 33046321, 33046322, 33449612, 33449712, 33449723, 33449817
- Models: CoaguChek XS PT Test 2x24 Strips, CoaguChek XS PT Test 6 Strips, CoaguChek XS Test 24 Tests USA
- Manufacturing and Distribution Dates: January 12, 2018 – October 29, 2018
- Devices Recalled in the U.S.: More than 1.1 million packages
The lot number is printed on the test strip label, which is applied to the test strip box and the test strip vial.
Device Use
CoaguChek XS PT Test Strips are used with Roche INR Test Meters to monitor patient response to warfarin, a blood thinner prescribed to prevent and treat blood clots. The test strip is inserted into the test meter, and then a medical pricking needle (a lancet) is used to obtain blood which is applied to the test strip. The meter reads the test strip, measures how long it takes the blood to clot, and provides the result based on a standardized calculation in the form of the International Normalized Ratio, or INR. The INR is used by patients and physicians to determine if warfarin dosing is appropriate.
Reason for Recall
Roche Diagnostics, the manufacturer of CoaguChek meters and test strips, is recalling the CoaguChek XS PT Test Strips due to inaccurate INR test results, when compared to laboratory results. Roche re-calibrated the CoaguChek XS PT Test Strips in January 2018 to correspond to a newly released INR International Standard. Since this re-calibration, Roche Diagnostics has received reports of patients experiencing abnormally high or inaccurate INR test results when testing with the affected CoaguChek XS PT Test Strips listed in the table below.
Use of affected products may increase the risk of serious adverse health consequences, including death.
Who May be Affected
All patients who have been prescribed the blood thinner warfarin (also known by the brand names Coumadin and Jantoven) to prevent and treat blood clots.
Patients with the following conditions are at especially high risk for serious events associated with inaccurate INR measurements:
- Mechanical heart valve
- Atrial fibrillation and high-risk CHA2DS2-VASc scores
- Recent thromboembolic events
What to Do
- Patients and health care providers should switch to new batches of test strips that are calibrated to the previous international standard, which Roche Diagnostics will provide to customers within one month.
- Patients should also contact their patient self-testing service providers to find out when they will be receiving corrected test strips
- Patients who are using CoaguChek meters and CoaguChek XS PT Test Strips affected by the recall should contact their health care provider and patient self-testing service providers immediately to determine alternative test methods and address questions regarding their testing schedule.
- Patients should consult with their health care provider before making any changes to their warfarin dose.
- Health care providers and patients with questions may contact Roche Diagnostics at 1-800-428-4674 to learn more details about the recall.
Contact Information
Roche Diagnostics Corporation
9115 Hague Road
Indianapolis, IN 46250-1025
1-800-428-4674
| Product | Catalog Numbers | Affected Lot Numbers |
CoaguChek XS PT Test 2x24 Strips | 04625315160 | 28124111 28124121 28631911 28631921 28631924 28632021 28632213 28632312 28632412 29415113 29415123 29494221 29494312 29494613 29494711 29778721 29779012 29779213 29779214 30497213 30497311 30497413 30497423 30497515 31404314 31404821 32264116 32264212 32264316 32264317 32264411 32264421 33045913 33046011 33046113 33046312 33046314 33046321 33046322 33449612 33449712 33449723 33449817 |
CoaguChek XS PT Test 6 Strips | 04625374160 | |
CoaguChek XS Test 24 Tests USA | 07797826160 |
Date Recall Initiated:
September 13, 2018
Additional Resources:
How do I report a problem?
Health care professionals and consumers may report adverse reactions or quality problems they experienced using these devices to MedWatch: The FDA Safety Information and Adverse Event Reporting Program either online, by regular mail or by FAX to 1-800-FDA-0178.