MRI Patient Monitor systemINVIVO MAGNITUDE MRI Vital Signs Monitor provides the operator with ECG, Respiration, SpO2, NIBP,EtCO2 and (Optional) Invasive Pressure monitoring. Full control of patient monitoring is provided via the INVIVO Millenia 3155A/3155MVS Monitor equipped with MRI shielding and set to Remote Communication Mode.
Every feature of the 3150M is synchronized via wireless 2.4 Ghz technology with the remote monitor which means this unit is portable and requires no expensive installation.
This system includes fiberoptic finger sensor, ECG Patient Leads and NIBP cuff and hoses.
FORUMSView All (6)
1Reply10 years ago | LCD adjustments password Anyone know this password? Reply |
1Reply11 years ago | invivo mri 3150 monitor how does it operate? Reply |
3Replies12 years ago | 3150`s for sale I have 3 3150m pulse oxymeters for sale--make an offer--Dave Burke 225-291-2054/P>Reply |
Among other features the Magnitude 3150M MRI Monitor includes:
"Digital Signal Processing" (DSP)
DSP provides improved ECG performance and reduction of MRI gradient artifact.
"MRI View Mode"
The "MRI View Mode" provides Adaptive ECG Gradient Filtering techniques for enhanced ECG display during MRI sequences.
VITAL SIGN PARAMETERS
ECG, Heart Rate, SpO2, NIBP, IBP, Temperature,
Anesthetic Agents, O2, EtCO2, N2O
ALARMS
Alarm Types: Medical and technical
Alarm Actions: Audible and visual alarms, configurable
automatic recorder printouts
Audible Alarms: Tones and bleeps, adjustable volume
Visual Alarms: Flashing red numeric, red waveform,
text message for technical alarms
Alarm Mode: Fixed and flexible
Alarm Silence Period: 3 min audible silence with visual
alarm indication
ECG
Heart Rate Range: 30-250 BPM
Lead Selection: I, II, III, aVL, aVF, aVR
Measurement Accuracy: ±0.5% of display, ± 1 BPM
Display Resolution: 1 BPM
Frequency Response: Monitoring Mode: 0.5 to 40 Hz
MRI Gradient Filtering: Monitor Mode: Digital Signal
Processing (DSP)
MRI View Mode: Adaptive Filter DSP
MR2: CVMRI gradient filter
MR3: CVMRI gradient with Adaptive Filter
MR4: T-wave suppression for high field MRI
Sensitivity: Adult >200µV
Neonatal >100 µV
Isolation: Meets IEC 60601-1 and UL 544 standards
Input: ECG via 4 lead patient cable
Defibrillation Protection: Up to 360 J, at 5 kV, recovery
within 1 second
Patient Leakage Current: 104dB
Input Impedance: >8 Megohm at 10 Hz
ECG ALARMS
Alarm Limit Range: 30 to 249 BPM
PULSE OXIMETRY
Measurement Range: 50 to 100%
Measurement Type: Red & Infrared light absorption
Sensors: Reusable Grip Sensors and Oxiclip
Grip Sensor Sizes: Adult, Pediatric, Infant, Neonatal
Accuracy: 100 to 90% ±2.0%
90 to 80% ± 2.2%
80 to 70% ± 2.5%
70 to 60% ± 2.8%
60 to 50% ± 3.3%
MRI ROOM POSITIONING
3150M: 5,000 Gauss (0.5 T) field line
3155A/3155MVS: 1,000 Gauss (0.1 T) field line
DISPLAY
Screen Size: 10.4 inches (26.7 cm) diagonal
Screen Type: Active Color LCD
Resolution: 640 x 480 pixels
WAVEFORMS & NUMERICS
Number of Channels: 4 with associated numerics
Numerics: 8 parameter selections, showing
real time values, units, alarm limits
& other relevant information
Display Layout: User configurable, 7 customized
layouts can be permanently stored
Single Trace Height: 20 mm
Numeric Character Height: Up to 25 mm (HR)
Sweep Speed: 25 or 50 mm/s for ECG, SpO2, IBP
Waveform Length: Approx. 6.7 sec at 25 mm/sec
TRENDS
Graphical Trends: All monitored parameters
Time Viewing: 20 min, 2, 4, 8, 24 hours
Trend Cursor: Time scale zooming
Tabular Trends: NIBP, HR, SpO2, CO2, Resp
Documentation: Integrated dual channel recorder
CARDIAC AND PERIPHERAL GATING
Interfaces for GE, Hitachi, Philips, Siemens, and Toshiba
Method: Digital
Delay: 280 mmHg or
after 3 minutes
NON-INVASIVE BLOOD PRESSURE ALARMS
Alarm Limit Range: 5 to 249 mmHg
Number of Channels: 2 simultaneous channels
Measuring Range: -10 to 250 mmHg
Accuracy: ±0.5% of reading, or 1 mmHg
(whichever is greater)
Frequency Response: O to 12 Hz
Zero Adjust: Zeroes with 300 mmHg offset to
0 ± 1 mmHg within 1 second
Transducer Requirement: 5µV/V/mmHg
Digital Pressure Display: Systolic/diastolic and mean
Alarm Limit Range: -10 to 250mmHg
FiberTEMP TEMPERATURE
Channel: One
Scales: Celsius or Fahrenheit
Range: 20.0° to 44.0°C (68.0° to 111.2°F)
Accuracy: +/- .3°C (.5°F)
TEMPERATURE ALARMS
Alarm Limit Range: 20.0° to 44.0°C (68.0° to 111.2°F)
ANESTHETIC AGENTS
Measurement Technique: Side Stream
Measurement Method: Automatic 5 Anesthetic Agent Identification
Technique: Non-dispersive Infrared Absorption
Display Update Interval: Breath by breath waveform
Calibration: No field calibration required
Measurement Range: Halothane 0.15 to 8.5Vol.%
Enflurane 0.15 to 10 Vol.%
Isoflurane 0.15 to 8.5Vol.%
Sevoflurane 0.15 to 10 Vol.%
Desflurane 0.15 to 22 Vol.%
Carbon Dioxide 0 to 10 Vol.%
Nitrous Oxide 0 to 99 Vol.%
Oxygen 0 to 100%
Accuracy: 5 Agents ±0.15 Vol.% + 15% relative
CO2 ±4mmHg or 12% relative
(whichever greater)
N2O ±2% Vol. + 8% relative
O2 ±3%, Full Scale
CARBON DIOXIDE (END-TIDAL CO2)
Measurement Technique: Side Stream EtC02 and N2O
Measurement Range: CO2 O to 76 mmHg
N2O 0 to 99%
Measurement Rise Time: < 300 msec at 230 ml/min. sample rate
Accuracy: CO2 ±3 mmHg
N2O 5% of full scale