Stryker - TP600
Heat therapy unit
The GAYMAR T/Pump® Localized Temperature Therapy System provides therapy by warming or cooling the enclosed water, and circulating it through the Gaymar Mul•T•Pad. The pad is connected to the Gaymar T/Pump with easytouse Clik•Tite® or Colder® -style connectors.
- Electrical Classification: Class I equipment with Type BF applied part suitable for continuous operation. Not classified for protection against ingress of liquid. Not classified for use in the presence of flammable anesthetics.
- Size (approx.) 11.5” x 8” x 8” 29.2cm x 20.3cm x 20.3cm
- Weight 6.5 lbs (2.9 kg) when empty 9 lbs (4.0 kg) with unit filled with water to heating level
- Reservoir capacity 93 oz (2.75 l) maximum
- Flow rate 9 gph (34 lph) minimum with pad attached
- Ambient operating temperature 60°F to 90°F (15.6°C to 32.2°C )
- Environmental conditions for transport and storage -20°F to 120°F (-28°C to 48°C ) At uncontrolled RH
- Temperature setpoints TP650 Classic 107°F (42°C) High heat 100°F (38°C) - Low heat 50°F (10°C) - Cooling TP700 Professional 107°F (42°C) 100°F (38°C) 95°F (35°C) 50°F (10°C) TP702 42ºC 38ºC 35ºC 10ºC
- Average temperature accuracy ±2°F at 107°F (±1°C at 42°C)
- Maximum Contact Surface Temperature 107°F (42°C)
- High Limit Safety Temperature 111°F to 118°F (44°C to 48°C)
- Power cord International (harmonized) 3-wire cordset Modular
- Current leakage 300 microamperes maximum 500µA Max
- Ground resistance 0.5 ohm max
- Electrical requirements Voltage (VAC) Frequency (Hz) Current (amps) 120±10% 60 3.1 amperes 230+ 10% 50 Hz 1.6 A
- Certifications MEDICAL ELECTRICAL EQUIPMENT WITH RESPECT TO ELECTRICAL SHOCK, FIRE, AND MECHANICAL HAZARDS ONLY IN ACCORDANCE WITH UL 60601-1, AND CAN/CSA C22.2 NO 601.1, ASTM F 2196-2002