2

Replies
-Job Hagiave
Wed Sep 13 2023
Biological Safety Cabinet Internal Blower is not turning on

Hi Support,

I have a problem with my Biological Safety Cabinet.

The internal blower is not turning on. Everything is all installed and fine except that the damper actuator is missing from the installation list. Blower fuse is OK. Connections are OK.

What can be the cause.

Cytotoxic cabinet details; NUAIR, Model:NU-430-400E, Class II B2, LABGARD ES

Can anyone help me with some expert technical advice to solve my problem please.


Thanks

Equipment: Biological Safety Cabinet - NUAIR NU-430-400E

Reply

FEATURES

  • Exhaust interlock
  • TouchLink2 electronic control
  • Intelli airflow control
  • For non-flammable/explosive chemicals
  • 100% exhaust

Additional Specifications

- Electrical data: 115V, 60 AC, 50/ 60 Hz

- Amps:  5 Blower (3 Lights)

- Amps::  3, (3) Duplex

- Amps:  8, (6) Total

- Concurrent Balance Valuw:  829CFM (1409 CMH)

- Power cord: 

Length: 12 ft.

14 GA - 3 Wire, 15A

- Certification Exhaust Value:  754 CFM (1281 CMH)

- Plant Duct Static Pressure: 1.7” w.g. (43 mm w.g. g.)

- Heat Rejected, BTU, Per Hour:  474

- Weight: 520 lbs. (236 kg.)