STERIS - Reliance 333 Series
by STERIS
For automated cleaning, thermal disinfection, and drying of reusable general surgical instruments prior to sterilization in outpatient/same-day surgery departments, surgical centers, and ambulatory care centers. A wide variety of surgical instrumentation and containers may safely be processed in the Amsco Reliance 333 including: reusable, general surgical instruments, minimally invasive surgery (MIS) rigid instruments, and microsurgical instruments. The Reliance 333 Washer/Disinfector is a single-door, counter height washer/disinfector, designed for installation under-counter or as a freestanding unit (with accessory panels). The standard pull-down door facilitates loading of the unit. An optional load lifting door is available to adjust the door height to countertop level for ergonomic ease of loading or unloading. The washer/disinfector features state-of-the-art microcomputer control, with user-friendly operator interface. A printer can be installed to record and monitor cycles and cycle parameters.
0RepliesMon May 04 2026 | Steris 333 washer, error 12, sump rtd heating malf ive got a unit that worked fine for a month or so and now has this error. From what I can tell, the contactors(CL and CH) are never getting power to the coil. Ive checked the 2 fuses and drivers b/w main pcb and i/o but now i'm stuck and the schematics just arent thorough enough. Maybe a bad relay? But 1 schematic shows that same ssr sharing chamber heating and 1 of the chemical pumps, but that pump is working. Reply |
2RepliesTue Apr 20 2021 | Failed RTD with conjoining sump issues I Have a Steris 333 in the field that is throwing a fault 12 followed by a fault 7 upon canceling the cycle. i put a brand new sump pump in the machine and checked all the cold water flow valves. the only thing i haven't done is calibrated the RTD again. would replacing the RTD probes circumvent having to recalibrate the machine? if not what's the best way to go about recalibration.Reply |
1ReplyTue Oct 08 2019 | Reset Procedure for Maintenance Reminders? We need the reset procedure to clear the Maintenance Code #3 reminder. Steris will not provide this information and says we have to purchase the manual. They will sell the PMs kits, but will not tell you how to reset reminders. Reply |
[list] [*] Cold water pre-wash [*] Enzyme wash utilizing a pulsed spraying pattern for effective protein removal [*] Neutral detergent wash with bi-directional spray arms on top and bottom of each load level [*] Thorough rinsing of instruments [*] Guaranteed thermal disinfection rinse at minimum 180° F for 1 minute [*] Instrument lubricant added during thermal rinse [*] HEPA Filter drying of complete load [/list]
Exterior dimensions: • 38 x 34* x 29" (965 x 864 x 736 mm) *35-1/2" (903 mm) high with countertop Interior wash chamber dimensions: • 24 x 19 x 23" (610 x 482 x 585 mm) Load height: • With manual (pull-down) door: (11-12" (292 mm) • Optional load lifting door, lowered: 11-12" (292 mm) • Optional load lifting door, raised: 27-1/4" (692 mm)