Zimmer - ATS 1500
Tourniquet
The Zimmer ATS 1500 is an automatic tourniquet which has been found useful in producing a bloodless operating field in a surgical procedure involving extremities. It is indicated to temporally occlude blood flow in patient's extremities during a surgical procedure on those extremities.
FORUMSView All (18)
2Replies5 years ago | Error code CAL 2 FAIL Zimmer - ATS 1500 showed at display Error code CAL 2 FAIL what should i do to fix this problem Reply |
3Replies7 years ago | gialgaca10@hotmail.com Customer reported that unit would not power up. Charged for several hours. Still would not. Battery voltage ok. Fuse ok. Was about to give up for the day when suddenly the unit started going through a series of self tests, then eventually displayed "ROM Fail" accompanied by a loud, long beep. Reply |
1Reply7 years ago | HELP !!!! Customer reported that unit would not power up. Charged for several hours. Still would not. Battery voltage ok. Fuse ok. Was about to give up for the day when suddenly the unit started going through a series of self tests, then eventually displayed "ROM Fail" accompanied by a loud, long beep. Reply |
- Indicated to temporally occlude blood flow in a patient’s extremities during surgical procedures on those extremities
- Tourniquets have been found useful in producing a bloodless operating field in surgical procedures involving the extremities.
- Line Voltage Rage: 105-130 VAC, 50/60 Hz
- Line Current: 485 mA RMS Max., 300 mA RMS nominal
- Input Power: 58 W max, 36 W nominal
- Power Cord Type: SJT, AWG 18, 13 ft. 2 in (4.01m) Power Plug Hospital grade, 3 prong straight blade, 15A
- Line Protection: 0.5A, 250V, 5 x 20 mm SLO-BLO fuse
- Cuff Pressure Range: 50-475 mmHg
- Pressure Accuracy: ± 5 mmHg (50-475 mmHg)
- Pressure Regulation: ± 6mm of set point
- Maximum Pressure: 475 mmHg (Nominal operation), 600 mmHg (internally limited)
- Time Alarm Set Ranges: 0-240 minutes ; 1 minute increments
- Timer Accuracy: 0.25% ± 1 second
- Height: 16.13 in (41 cm)
- Width: 8.75 in (22.3 cm)
- Depth: 7.1 in ( 17.9 cm) (Including 2 in clamp)
- Weight: 19.8 lbs (9 kg)