GE HealthCare - Dash 2500
The Dash 2500 monitor from GE Healthcare is the latest vital signs monitor in the Dash Family.The Dash 2500 monitor from GE Healthcare is the latest vital signs monitor in the Dash Family.
The Dash 2500 is a reliable and affordable, bedside or mobile monitor that provides the clinical information you need to assess and treat your patients with speed and accuracy.
By combining Dash family performance and Gold Standard Dinamap Superstat NIBP technology, the Dash2500 fits all your needs in every hospital department. It allows you to deliver a new standard of clinical excellence to all patient groups in low-acuity settings such as ER, recovery/PACU, outpatient surgery, GI/endoscopy, progressive care, interventional radiology, general care and special care units.
The Dash 2500 includes sophisticated clinical parameters to capture patient vital signs and unparalleled cardiac monitoring to help accurately detect arrhythmias.
An extensive feature set enables you to create a portable monitoring solution capable of meeting a wide range of demands. All the features, you expect, are included as standard.
FORUMSView All (5)
2Replies9 months ago | NIBP pump timeout The piece that connected the hoses with the main board is missing and i need a alternative solution for this i tried to contact the tube directly that what i get (NIBP pump timeout) It rech 180 and land to 60 without results Equipment: Ge Dash 2500 patient monitor Reply |
4Replies4 years ago | Loss of memory in patient monitors When patient monitors displays loss of memory which particular component is dead or weak ?Reply |
1Reply6 years ago | Password for GE Dash 2500 Hello, I need the password the access on the config and service menu of the GE Dash 2500. Can you send me it? Thank youReply |
- 3 and 5 lead ECG -> EK-Pro™ arrhythmia analysis
- Respiration monitoring
- Non-invasive blood pressure -> DINAMAP® SuperSTAT™, The Gold Standard for NIBP accuracy
- Pulse Oxymetry -> Masimo SET® or Nellcor® OxiMax® SpO2
- Temperature: Alaris® Turbo Temp®
Height | 8.7 in |
Length | 6.7 in |
Weight | 12 lbs |
Temperature Range | 5 to 40 °C |
Width | 14.1 in |
Heart Rate/Pulse:
Electrocardiography (ECG):
Heart rate accuracy 30 to 300 bpm, ± 3 bpm or 3% of reading, whichever is greater
Nellcor SpO2:
Range 20 to 250 bpm
Accuracy and tolerance 20 to 250 bpm ± 3 digits
Low perfusion 20 to 250 bpm ± 3 digits
Masimo SpO2:
Range 25 to 240 bpm
Accuracy and motion tolerance
Without motion 25 to 240 bpm ± 3 digits
With motion normal physiologic range,
25 to 240 bpm ± 5 digits
Non-invasive blood pressure (NIBP):
Adult/pediatric range 30 to 200 bpm
Neonate range 30 to 220 bpm
Accuracy ± 3.5% or 3bpm, whichever is greater
ECG:
Leads available:
3-electrode configuration I, II, III
5-electrode configuration I, II, III, aVR, aVL, aVF, and VA
Heart rate resolution 1 bpm
Bandwidth 0.5 to 40 Hz +1/-6 dB, 0.05 to 40 Hz +1/-6 dB, 0.05 to 100 Hz +1/-6 dB
Standardizing voltage 1 mV marker
Common mode rejection 1 mV RTI or 10 mm p-p max displayed noise allowed with 20 Vrms, 50-60 Hz input
Input impedance
Common mode > 2.5 MΩ at 10 Hz
Differential > 2.5 MΩ from DC to 60 Hz
60 Hz tolerance up to 10 mV
Pacemaker detection/rejection
Input voltage range ± 2 to ± 700 mV
Lead off sensing current < 0.1 ΜA DC signal leads, < 1 ΜA DC driven lead
Respiratory:
ECG-Derived respiration rate
Leads available I or II
Range 6 to 120 breaths/min (adult/pediatric), 6 to 180 breaths/min (neonate)
Accuracy ± 2 breaths/min or ± 3% of reading, whichever is greater
Resolution 1 breath/min
Base impedance 100 to 2000 Ω
Detection sensitivity 0.2 Ω at 30 breath/min with 500 Ω baseline impedance
BP Measurement ranges:
Systolic:
30 to 290 mmHg (adult/pediatric)
4.0 to 38.7 kPa (adult/pediatric)
30 to 140 mmHg (neonate)
4.0 to 18.7 kPa (neonate)
MAP:
20 to 260 mmHg (adult/pediatric)
2.7 to 34.7 kPa (adult/pediatric)
20 to 125 mmHg (neonate)
2.7 to 16.7 kPa (neonate)
Diastolic:
10 to 220 mmHg (adult/pediatric)
1.3 to 29.3 kPa (adult/pediatric)
10 to 110 mmHg (neonate)
1.3 to 14.7 kPa (neonate)
Resolution: 1 mmHg
Accuracy: Meets AAMI/ANSI standard SP10:2002
Initial cuff inflation pressure: 135 ± 15 mmHg default; user selectable (adult/pediatric)
100 ± 15 mmHg default; user selectable (neonate)
Maximum determination: 120s (adult/pediatric) time 85s (neonate)
Over pressure monitor: 300 to 330 mmHg (adult/pediatric) 150 to 165 mmHg (neonate)
Hose/cuff interface: Compatible with current Dash hoses
Pulse rate: When NIBP is the source, HR values are derived from the pulse rate that is determined by the oscillometric technique of measuring blood pressure. The rate source field is labeled NIBP.
Adult/pediatric range: 30 to 200 bpm (± 3.5% or 3 bpm)
Neonate range: 30 to 220 bpm (± 3.5% or 3 bpm)
Nellcor OxiMax SpO2:
Measurement range
SpO2 1 to 100%
Pulse rate 20 to 250 bpm
Accuracy
Saturation
Adult 70 to 100% ±2 digits
Neonate 70 to 100% ±3 digits
Low perfusion 70 to 100% ±2 digits
Pulse Rate
Adult and neonate 20 to 250 bpm ±3 digits
Low perfusion 20 to 250 bpm ±3 digits
Masimo SET SpO2:
Measurement range
SpO2 1 to 100%
Pulse rate 25 to 240 bpm
Saturation
Without motion– 70 to 100% ±2 digits adult/pediatric
Without motion– 70 to 100% ±3 digits neonate
With motion– 70 to 100% ±3 digits adult/pediatric/neo
Low perfusion 70 to 100% ±2 digits, 0 to 69% unspecified
Pulse Rate
Without motion 25 to 240 bpm ±3 digits
With motion normal physiologic range 25 to 240 bpm ±5 digits