Medtronic - Physio Control Lifepak 10
by Medtronic
Defibrillator
LIFEPAK 10 defibrillator/ monitor/pacemaker was designed to withstand the harsh environment faced by emergency medical providers throughout the world.
FORUMSView All (9)
-Azin
Tue Jan 18 2022
Service Error
The "service" alarm is on How can I fixe that?Reply
-Chris-CBET
Mon Aug 14 2017
Testing issues
While performing the semi-annual inspection of a customer's defib, I noticed blank recording and internal discharge results in all characters on display. The VSP's show use, they have no continuity, but i can see carbon residue(blackened) on most of them(VSP2 - 6). Has anyone seen this? Bad inductor for the paddle tray, and bad capacitors for the recorderReply
2
-Dimitris Christidis
Fri Apr 21 2017
Ac power module
Can lifepak 10 operate properly only with the AC power module since the batteries have been depleted??Reply
  • Primarily used in Ambulances and Emergency Rooms
  • Lightweight, portable defibrillator monitor
  • Built-in Recorder
  • Battery operated, AC adapter optional
  • Requires separate battery charging system
  • Meets current ACLS standards
  • Easy-to-read display
Weight20 lbs
Defibrillator TypeExternal, Automated (AED)
  • Isolated ECG via quick-Look defibrillator paddles, Fast Patch electrodes, or 3-lead patient cable
  • Display Size: 72.55 mm (2.85 in)  x 43.5 mm (1.7 in), non-fade
  • Charge Time: Charges to 360 joules in less than 12 seconds above 0°c
  • Paddle Area:82 cm2