Valleylab - Force 2
by Valleylab
Provides cutting and coagulation in monopolar or bipolar outputs.
The Force 2 electrosurgical unit from Valleylab provides cutting and coagulation in monopolar or bipolar outputs. Force 2s have audible and visual alarms, including REM™ Fault Indicator. The REM™ (Return Electrode Monitoring) is standard. 300 watts is available for output in the cut setting. A monopolar foot switch is included, a bipolar foot switch is optional as is a rolling cart.
FORUMSView All (50)
-Chris-CBET
Thu Jun 25 2026
Valleylab force 2-2PCH
So i'm trying to calibrate the REM, following instructions from the 2004 service manual(not the 1990 with a different REM PCBA and numbered pots. If you've done this before you can see how half-assed these instructions are(and imagine how much i regret this now))anyway, it says to turn R11 for the highest voltage(no mention of direction, ac or dc nor where you're attaching.(I ended up attaching my leads(neg) to the TPgnd1 and (pos)to TP3 after hunting down EREM on a schematic, and monitoring DC) next turn r15 counter-clockwise till it's below 4v. Except when i turn counter-clockwise the voltage only goes up, if I try to turn clockwise it goes down, so i get it below 4v, next i move my lead from tp3 to tp5 and i have no voltage reading now, so how am i supposed to turn r15 tills it's above 5? What am I doing wrong(other than throwing it in the trash and selling them something newer
-MJ-925
Mon Apr 24 2023
100 W Output limit for Force 2
This Force 2 (Dash number 20CEM) will not output any monopolar power higher than 100 Watts. The power settings can be adjusted to the maximum values for each mode on the display (300 - pure cut, 250 - Blend 1, 200 - Blend 2, 150 - Blend 3, 120 - Coag). But when the footpedal or handswitch is activated, the unit will reset the power setting to 100 W, then output power at that setting. The Bipolar mode does not seem to be affected by this. This seems like a setting that needs to be adjusted. Any input is appreciated.
-IBMA
Sat Dec 31 2022
High Frequency Current Leakage test above the limit 150µA
Hello Guys, I have a Valleylab force2 which works well but when I did the high frequency current leakage test, it did not pass the coagulation tests. It gives me values ??above the limit 150µA If something similar has happened to you, it would help me to know what solution you gave it, thanks.
Equipment: Valleylab - Force 2
BLENDED CUT MODESProvide flexibility through varying degrees of hemostasis. Choose one of three preset modes.                        SIMULTANEOUS INDEPENDENT COAGULATIONPermits two surgeons to fulgurate from a single generator, for added convenience and efficacy. Accessories are activated only when keyed, reducing the potential for injury caused by inadvertent activation.                        LOW VOLTAGE COAGULATIONEnsures controlled, precise desiccation with less destruction of peripheral tissue, making the Forceª 2 generator ideal for laparoscopic procedures.                         REM™ SAFETYIs guaranteed with Valleylab's patented adaptive REMª system, which continually monitors patient impedance levels. If a fault in the patient/return electrode contact is detected, the REMª system automatically deactivates the generator virtually eliminating the risk of burns under the return electrode. Valleylab REMª safety has been proven in more than 100 million surgical procedures worldwide.
Height8 in
Length 21 in
Weight23 lbs
OutPut Frequency500 kHz
Width13 in