Fluke Biomedical - DALE 601E
by Fluke Biomedical

Follow Us
DESCRIPTION
Electrical Safety Analyzer 230 V

The DALE601E Electrical Safety Analyzer is a trusted tool for biomedical engineers, offering an extensive range of electrical safety and performance measurements for medical equipment. Its intuitive single master function switch guides users step-by-step through complex testing procedures, ensuring efficient and accurate diagnostics.

The DALE601E accurately measures line voltage, instrument current, ground resistance, earth and chassis leakage, patient lead-to-ground, lead-to-lead isolation, M.A.P. current, external gradient, and transducer isolation current, among others. Designed for precision, it incorporates a four-wire Kelvin bridge to eliminate resistance reading errors due to cable length and contact resistance.

Featuring true RMS current readings, a selectable AAMI or IEC test load, and five universal EKG lead connectors, this analyzer supports virtually all lead types in clinical and laboratory settings. Heavy-duty switches simplify polarity changes, neutral opens/closes, and leakage current test selection. (Generated by AI)

FORUMSView All (2)
Ask a New Question

3

Replies
-Gruven
Sat Jul 13 2013
used dale 601e

anyone know where I can find a decent priced Dale 601e. Fluke wants just way too much for this item. 

 

Thanks,

Nick Putch

Reply

2

Replies
-Jack F
Tue Apr 30 2013
Fluke does not pass calibration testing

I have a question concerning a dale 601e meter we sent in for calibration testing at a calibration facility :  The test facility measured a high resistance between earth gnd. of AC inlet and output test receptacle of 34 kOhm.   Does that measurement  have a specification.  I did not find one in the manual.

Thanks

Jack

 

 

Reply
DOCUMENTS / MANUALSView All

FEATURES

[list]          [*] Small, lightweight, self-contained portable instrument          [*] Master range meter delivers accurate measurement for each test          [*] Accurate test measurements with 4-wire Kelvin bridge          [*] Leakage Current can be measured in Open/Closed Neutral, Normal/Reversed Polarity and Open/Closed Ground          [*] All Leakage Current measurements are displayed directly in µAmps          [*] Broad range of external measurements          [*] Five (5) Universal EKG Lead Connectors          [*] The DALE601E is designed for 230 V operation          [/list]

SPECIFICATIONS

Height196 mm
Length 41 mm
Analyzer/Simulator TypeElectrical Safety
Current Range0 to 10 A
Width102 mm
This site uses cookies. By continuing to browse the site you are agreeing to our use of cookies. Please review our Privacy Policy for more details.