Fluke Biomedical - DALE601
by Fluke Biomedical
The DALE601 measures Line Voltage, Instrument Current, Ground Resistance, Earth and Chassis Leakage Current, Patient Lead-to-Ground, Lead-to-Lead and Lead Isolation (M.A.P.) Current, External Gradient, Probe and Transducer Isolation Current and Point-to-Point Resistance.            DALE601 also tests all 120-volt standard wall outlets for proper wiring. The green and amber light indicate five normal or fault conditions. Unique to the DALE601is its capability to make a broad range of external measurements. A clearly labeled single master function switch guides the user through a complete measurement procedure.            Heavy-duty switches change polarity, open and close the neutral and select Chassis or Earth Leakage Current measurements. A four-wire Kelvin bridge eliminates reading errors in resistance due to cable length and contact resistance. True RMS measurements are provided for all current readings. The AAMI test load is utilized.            The DALE601 has five (5) universal EKG lead connectors that allow for any type of lead connection.There is also a single slide switch that gives the user a choice of AAMI or IEC test load.
-ereynolds
Tue Feb 22 2022
2A Fuse
Hi, the DALE601 im currently working on is repeatedly blowing its 2A fuse. Wondering if anyone has any suggestions on what could be causing this.Reply
-Vetsens
Tue Sep 13 2016
Service Manual
Hi, we are looking for the service manual for he Dale Technology 600/601 safety analyser. Specifically, we have several unit which have broken teansformers from being dropped and would like to fix these units. Thanks please email info@vetsens.com
[list]      [*] Small, lightweight, self-contained portable instrument      [*] Master range meter delivers accurate measurement for each test      [*] Accurate test measurements with 4-wire Kelvin bridge      [*] Test method complies with AAMI load per ANSI/AAMI ES1-1993 or IEC601-1      [*] Leakage Current can be measured in Open/Closed Neutral, Normal/Reversed Polarity and Open/Closed Ground      [*] All Leakage Current measurements are displayed directly in µAmps.      [*] Broad range of external measurements      [*] Five (5) Universal EKG Lead Connectors      [*] The DALE601 is designed for 120 V operation      [/list]
Height7.8 in
Length 1.6 in
Analyzer/Simulator TypeElectrical Safety
Current Range0 to 19.99 A
Width4 in