Siemens - Polymobil
by Siemens
Mobile C-Arm
POLYMOBIL makes it easy to bring the X-ray system to the patient. They are as easy to use as they are to move, and offer good reproducibility and results in consistent image quality. The systems of the POLYMOBIL family are ideally suited for any bedside exposure in hospital ward, trauma practice, surgery and orthopedics and veterinary medicine. [b][i]Not commercially available in the United States (U.S.A.)[/i][/b]
FORUMSView All (31)
-FernandoGaro
Tue Feb 25 2025
Error 6
Hello everyone.
I have a 2013 Polymobil Plus showing error 6.
I followed all the recommendations in the manuals I found on the web, but without success.
The D927 board that the equipment contains has several differences in relation to the few provided schematics.
The voltage of 450 VDC is present on the IGBTs.
When I took measurements with an oscilloscope at the points specified in the Service Manual, I didn't get any readings.
I noticed that the V46 LED, which indicates that the internal power supply is working, doesn't light up.
Where am I going wrong?
Thank you all.
Equipment: Polymobil Plus Year 2013 serial # 50909
-Ajin134
Wed Nov 20 2024
error 34
master in x-ray i need help how to fix error 34
hope you could help me
thanks in advance
Equipment: Siemens polymobil plus
-adell
Sat Mar 12 2022
Error 14
Hello everyone I have a problem when I click on the X-ray button, it gives me error 14, and I tried to search in service manual and I didn't find anything about error 14 Reply