Abbott - Cell-Dyn 3200
by Abbott
Hematology analyzer
The CELL-DYN 3200 is a multi-parameter automated hematology analyzer designed for in vitro diagnostic use in clinical laboratories. The analyzer uses optical light scatter for WBC, RBC, PLT and Reticulocyte analysis to meet the needs of today’s medium volume laboratories. The instrument offers simple operation, accurate 5 part WBC differential analysis, automated sample handling and reliability. The system combines MAPSS (Multi-Angle, Polarized Scatter Separation) technology with flow cytometry to achieve a high level of accuracy and precision.
FORUMSView All (3)
-TMPhD
a year ago
RBC overpressure
I am getting RBC syringe overpressure error. I followed the troubleshooting from the manual and have cleaned the diluent sheath and the main syringe. I can clear the fault, but then reoccurs when initializing. What else can I do?
-TMPhD
6 years ago
normally closed valves
the diagrams in the operators manual does not specify which tubing goes in the valve upon installation. service manual difficult to read. I am having backflow of reagents, assuming it is caused by misplaced tubing. Can someone share a photo how it looks in the 6 valves? Or describe which tubing should be in the valve?Reply
8
-LabDr
7 years ago
flow cell adjust
how do you adjust the flow cell for the differantials in a Cell Dyn 3200 Reply
  • Available as a single tube closed sampling system or 50 position sample loader for up to 71 samples per hour.
  • Data storage for 10,000 cycles with graphics for easy review of patient results.
  • Only 3 reagents are needed for complete CBC with 5 part differential analysis.
  • Dual angle, optical light scatter to enumerate and size platelets.
  • Accurate RBC counts and indices using 3 angles of light scatter and flow cytometry for accuracy.
Height19 in
Length 33 in
Weight232 lbs
Width23.5 in
  • Sample Volume: Open Mode 150 ΜL, Sample Loader 230 ΜL
  • Reagents: Only 4 reagents including reticulocytes
  • Platelets: Dual angle optical analysis, no extra reagent, no reflex testing requirement
  • Reticulocytes: New Methylene Blue NCCLS method, supravital staining technique
  • WBC: 0.02246.8 x 103/ΜL
  • RBC: 0.007.50 x 106/ΜL
  • HGB: 0.025.0 g/dL
  • HCT: 8.3-79.8%
  • MCV: 58139 fL
  • RDW: 10.0-29.8%
  • PLT: 0.03000 x 103/ΜL
  • MPV: 4.3-17.2 fL
  • RETC: 0.222.9%
  • Power Requirements: VAC 100-240V, 47/63Hz
  • Maximum Power: 550 watts