Sorvall - RC-5C
Superspeed Centrifuge.The RC-5C Plus offers simplicity of operation and durability for intensive use. User-friendly controls make set-up simple even for inexperienced users, speeding up parameter setting and reducing errors. The industry’s widest rotor range spanning fixed angle, swinging bucket, vertical, microtube, zonal, and continuous flow types means unrivalled levels of superspeed versatility.
FORUMSView All (5)
9Replies7 years ago | Temperature control I'm having intermittent failure on the temperature control. Sometimes it sticks on and cools the rotor to -18 degrees, sometimes during runs if fails to kick on the compressor and the temperature goes up to the limit and shuts the run down. Is there a relay that might be stuck or would this be a faulty temperature sensor? There are 2 sensors, one for air and one for the chamber temperature. Thank you.Reply |
0Replies8 years ago | Sorvall 5C-RC centrifuge with light A4 off A5 on and cold light on Does the above condititon pretty much gurarantee a tachometer problem. Thanks don Reply |
1Reply8 years ago | fault 7 Reply |
[list] [*] RCF mode for g force calculation [*] Microprocessor control [*] LED display of operating status [/list]